It is not uncommon to have “easter eggs” in electronic equipment with complex firmwares.
Often hidden functions are implemented for debug and development purposes, but sometimes just for run too.
Firmware engineers are human beings too, after all. In article below some of secret functions of Keithley test equipment are covered.
Feel free to try these on your gear, but do remember that information provided here is “AS IS” and no warranty provided by vendor or any other 3rd party. For example, changing Model ID on Keithley 24xx SourceMeter units cause permanent loss of calibration data. Obvious culprits are noted in this article, but something can be lost still.
Let’s get started.
Overview of how key pressing works when powering the meter (K2000, K2010 and alike).
Communication between main/digital board (MB) and front panel (FP) is done via simple UART, 9600 bps. Around 2.168 ms after reset line from MB become high, FP reads buttons state connected to PA3 MCU pin. This state is saved and then transmitted as part of FP identification message that FP sends as response for 0×0F command.
Here are such messages for K2001-style and K2000-style front panels respectively.
{'2', '0', '0', 'x', '/', '7', '0', '0', 'x', ' ', 'A', '0', '2', ' ', ' ', 0x80, 0x00}
{'2', '0', '0', '0', ' ', 'A', '0', '2', ' ', ' ', 0x80, 0x00}
Byte 0×80 near the end contains info about buttons. 0×80 – no buttons pressed. Other number represent sum of codes of buttons pressed – 7 bits available for
7 buttons.
Here is the table of button location for different units (according to schematics).
| MCU pin | Single button code | K2000/K2010 | K2001/K2002 | K7001 |
| PB0 | 0×81 = 0×80 + 0×01 | N/A | CAL | LEFT |
| PB1 | 0×82 = 0×80 + 0×02 | OPEN | FREQ | 6 |
| BP2 | 0×84 = 0×80 + 0×04 | CLOSE | MATH | 2 |
| PB3 | 0×88 = 0×80 + 0×08 | STEP | CONFIG | EXIT |
| PB4 | 0×90 = 0×80 + 0×10 | SCAN | DCI | OPEN |
| PB5 | 0xa0 = 0×80 + 0×20 | DIGITS | TRIG | STORE |
| PB6 | 0xc0 = 0×80 + 0×40 | RATE | INFO | CLEAR LIST |
Thus, pressing MATH + FREQ gives magic code 0×86. K2001-style panel used in other units such as K2400, K6517, … so buttons are on their places but with their own names.
Button actions for K2010, K2182 multimeter and similar single-line VFD display platform units.
| Code | Button | Action |
| 0×82 | OPEN | BUILT-IN and DIAG menu appear in TEST menu as they are in K2000, providing more tools for meter debugging and validation |
| 0×84 | CLOSE | Interactive mapping between all symbols that front panel can display and their hex codes |
| 0×88 | STEP | Temporarily loads default calibration constants, after reset meter loads as usual. Display will show “DEFAULT CAL” |
| 0×90 | SCAN | Loads DIAG menu immediately |
| 0xa0 | DIGITS | Displays all VFD symbols like during the start, waits for EXIT key to continue (also seen that if meter is turned off, all-on segments are still there after turning it on again if EXIT wasn’t pressed before …) |
| 0xc2 | OPEN + RATE | WARNING !!! Resets calibration constants to default ones, i.e. effectively erases all calibration data. That is irreversible action and there is no any confirmation before that, meter will just display “DEFCAL SAVED” |
Button actions for K2001, K2002 multimeters and similar with DUAL line VFD display.
| 0×81 | CAL | Enable low-level factory calibration |
| 0×82 | FREQ | Display advertisement message – meter capabilities, in English |
| 0×84 | MATH | Display advertisement message – meter capabilities, in French |
| 0×86 | MATH + FREQ | Enable secret menu |
| 0×88 | CONFIG | Display advertisement message – meter capabilities, in German |
| 0×90 | DCI | ??? Maybe something related to scan card |
| 0xa2 | TRIG + FREQ | Shows design team |
Now simple user representation of same information:
Keithley Model 2001 DMM (Axx firmware versions)
This 7½ multimeter features dual line dot matrix VFD. This kind of display is handy for displaying lots of text data.
1. Ensure unit is powered off. 2. Press and hold button "FREQ" and button "MATH" and power on unit 3. Secret menu available now in MENU/GENERAL/SECRET 4. After power down unit enters normal operation mode as usual.
Firmware versions with “A” index are more constrained in size due to only single 4Mbit ROM chip being used on digital board. It is a PLCC32 package 27C4096, located in socket near CPU. Here is information based on latest A08 version.
Secret menu have next items:
| Menu item | Detail | Default | Note |
| BKGD-DLY | Background delay | 5 | unknown effect |
| SIGNAL-DLY | Signal delay | 5 | unknown effect |
| CREDITS | Lists software team developers | ||
| SCANNER | Override scan card detection | No | Unlocks scanner card functions even without actual card present |
| INIT-BB-RAM | Initialize BB-RAM | Initialize option battery-backed NVRAM to operation data | |
| COMP-CAL | Skip AC in comprehensive calibration | No |
Init BB-RAM is a easy option to update memory options with replaced DALLAS NVRAM.
Before upgrade procedure was covered in article here which required use of ROM program tool.
Also when secret mode is enabled, there is extra 8½ resolution setting available in DCV function mode.
But its more for demo purposes, as 2001 analog board/ADC does not really provide necessary noise and linearity performance for real 8½ digit meters.
Software team using same digital board to develop functions for both 2001 and 2002 multi-meters, and were working on 2002 firmware even before actual 2002 analog hardware was available.
8½ mode is also available to use if use GPIB command 8.5;. By GPIB it’s possible to set ACV, DCI, ACV, 2wire Ohms, 4wire Ohms to higher 8½ resolution mode as well, by same :DIG 8.5; operand in
mode selection command, while front panel secret menu allows this only on DCV function.
Keithley Model 2001 DMM (Bxx firmware versions)
“B” version firmware is using larger ROM size, since it’s split between two ICs for odd and even bytes. One of ours 2001’s have digital board revision K, which have this configuration and using two ST M27C4001 DIP32 UV-erasable EPROM.
Here’s secret menu options with firmware version B15, kindly provided by one of Keithley Model 2001 owners.
Entrance is same as A-version 2001:
Ignore error codes, these are due to wrong calibration data in I2C EEPROM, since unit in video is under repair. Normal 2001 will not
have any error codes while entering secret mode.
| MESSAGES | Hidden messages submenu | Entry to show credit messages in english/french/german and game | |
| INIT-OPTIONS | Initialize EEPROMs for option cards | ||
| FREEZE | Freeze display | Off | Turning this on causes the PREV button to freeze display |
| OPTION | Override installed addon card detection | Enable meter to init EEPROM on option cards | |
| AC-CAL | Skip AC in comprehensive calibration | Allow manually view and change some of cal data, offsets/gains |
| MSGS Item | Detail | Note |
| CREDIT | Shows The Design Team: credits | Lists all team members who work on Model 2001. ~30 seconds |
| ENGLISH | Listing features and capabilities | English language demo mode (likely used for trade shows, ~2min) |
| FRENCH | Listing features and capabilities | French language demo mode ~ 2 minutes |
| GERMAN | Listing features and capabilities | German language demo mode ~ 2 minutes |
| GAME | Easter egg: game? | Entering this submenu on B15 firmware does nothing :) |
Also when secret mode is enabled, there is extra 8½ resolution setting available in DCV function mode, same as with A-version firmware.
Keithley Model 2002 DMM
Model 2002 have similar menu like 2001, but with more settings and more easter eggs in it.
This flagship Keithley DMM was designed with main competitor in mind, HP 3458A. It also have compatibility mode for GPIB interface,
so existing software could interface 2002 using 3458A SCPI commands.
Unlocking hidden menu is same as Model 2001:
Here is video overview of all items:
| MSGS | Hidden messages submenu | Entry to show credit messages in english/french/german and eggs | |
| INIT-OPTS | Initialize EEPROMs for option cards | ||
| OPTION | Override installed addon card detection | ||
| RAW-DISP | Display measurements in ADC counts | Shows counters data for ADC operation in various modes | |
| CAL-CONSTS | Setting offset and gains for DCV/Ohms | Allow manually view and change some of cal data, offsets/gains | |
| SKIP-AC-CAL | Skip AC in comprehensive calibration | ||
| BKGD-AV | Background averaging function | Enabled | Some digital filter, perhaps? |
Here details of second-level submenus:
| CREDIT | Shows The Design Team: credits | Lists all team members who work on Model 2002. ~30 seconds |
| HP-KILLER | Easter egg: 2002 eating 3548 digits | 2002 number are coming and eating 3548, funny demo :) |
| HP-SHOOTER | Easter egg: 2002 shoots 3548 digits | 2002 shooting rockets to 3548, killing one digit every time :) |
| INIT-OPTS | Detail | Note |
| BB-RAM | Initialize BB-RAM | Prepares DALLAS NVRAM to enable MEMx option installation |
| PREAMP | Initialize EEPROM on PREAMP | PREAMP option card module EEPROM initialize? Changes scan functions to PREAMP |
| PHASE | Initialize EEPROM on PHASE | PHASE measurement card module EEPROM initialize? Changes scan functions to PHASE |
| TC-OHMS | Initialize EEPROM on TC-OHMs | TC-OHMS option card module EEPROM initialize? Changes scan functions to TC-OHMS |
Keithley have Preamp option for 2001 and 2002, called Model 1801, and likely internal phase and TC-OHMs (thermo compensated ohms mode?) cards on modules which can fit Model 2002 in scanner card slot. Not sure last two do, but since enabling these options via secret menu changes scanner control functions and menus to related hidden menus, it’s the guess.
Preamp Keithley Model 1801 enhances low-side range measurement capability of 2001/2002 DMMs. There is scarce information available about it, just a brief sheet. According to which, preamp is designed to connect with DMM via option slot, instead of scanner card. According to datasheet, option card features just low-noise power supply to external Model 1801 nanovolt preamp module.
Gain with preamp module is 1000 on 2 mV range, 10000 on 200 uV range and 100000 on 20 uV range.
DMM with preamp option have only next ranges available:
| DCV | 20uV, 200uV, 2mV, no peak measurements, resolution down to 1 pV, Noise min 0.6 nV p-p, 1GOhm impedance |
| ACV | 500uV RMS low-frequency, no normal RMS/Peak/Crest factor measurements, resolution down to 100pV |
| 4W Ohms | 2 mOhm, 20 mOhm, 200 mOhm, 2 Ohm, 20 Ohm, 200 Ohm, resolution down to 100 pOhm, 9.2 mA current source, 5V |
| Frequency | 1 Hz to 1000 Hz, 5 digits resolution, 400uV p-p minimum signal level |
| Temperature | Thermocouple or uV/°C value, resolution 0.0001 °C) |
PREAMP configuration:
| PREAMP CONFIG | Detail | Note |
| CONTROL | PREAMP CONTROL: Enable or disable | Enable changes available ranges to those available with preamp |
| FILTER | Selects between filter speed of preamp | Available options Slow/Medium/Fast |
| CALIBRATION | Initialize meter calibration constants | Reads preamp calibration data and calibrate DMM to use them |
Phase card is another option. Google knows nothing about this item.
PHASE configuration:
| PHASE CONFIG | Detail |
| CONTROL | Phase measurement control: Enable or disable |
| SPEED | Measurement speed setting, like normal DMM functions in NPLC |
| FILTER | Filtering options, just like normal DMM functions |
| CHANNEL | Select phase options for signal or reference |
| SCALE | Setting phase scale between +-180 or 0-360 degree |
| UNITS | Setting degrees, radians or grads for phase unit |
| REF-RES | Setting reference resistance, in ohms |
| CAL | Calibrate phase reading |
TC-OHM did not change either settings, but I did not check every option yet.
Back to Secret menu, Options submenu:
| OPTIONS | Detail |
| NONE | No option override, detect by HW |
| SCANNER | Scanner card option override |
| PREAMP | Preamp card option override |
| PHASE | Phase measurement option override |
| TC-Ohm | Override option detect to TC-Ohm |
Keithley Model 24xx series SourceMeter
On 2400 series SMUs entering mode is using same buttons:
1. Ensure unit is powered off. 2. Press and hold gray button "^" and button "SWEEP" and power on unit 3. Secret menu available now in MENU/GENERAL/SECRET 4. After power down unit enters normal operation mode as usual.
Here are hidden menu items on Model 2400 with firmware C32:
| MODEL# | Override model ID | Allow to force ID to 2400/2410/2420/2430/2426/xxxx. Warning! Calibration will be LOST! | |
| FB-BURN-IN | Burn-in all VFD pixels | ||
| DIGITAL-REV | Write digital board revision ID into ROM | Stores digital board revision in calibration ROM | |
| ANALOG-REV | Write analog board revision ID into ROM | Stores digital board revision in calibration ROM | |
| CONTACT | Enable contact option 24xx-C(need hw) | Allow to enable CONTACT option hardware |
Of course, changing MODEL ID will not make SMU working with other ranges, since analog boards on different models within 24xx series are physically different.
This function is in firmware to allow developers test and use same digital board and same firmware, while connected to different analog boards.
Do not change Model# if you don’t have calibration-grade DMM available! Calibration data will be permamently lost!
Keithley Model 2510/2510-AT
1. Ensure unit is powered off. 2. Press and hold gray button "T" and button "CONFIG" and power on unit 3. Secret menu available now in MENU/GENERAL/SECRET 4. After power down unit enters normal operation mode as usual.
Hidden menu items on Model 2510 with firmware A11:
| BOARD-REV | Write board revision ID into ROM | Stores board revision in calibration ROM | |
| AUTOTUNE | Enable/disable autotune option | Allow to select between 2510 or 2510-AT Model SMU |
Keithley Model 6517A Electrometer
This unit does have secret menu as well, but since don’t have one (just yet?) unable to make video for it. Secret mode is also required to enable access to meter’s full diagnostics and calibration menus.
1. Ensure unit is powered off. 2. Press and hold button "SEQ" and button "Voltage source DOWN" and power on unit 3. Secret menu available now in MENU/GENERAL/SECRET 4. After power down unit enters normal operation mode as usual.
Firmware code C05 for 6517A“http://doc.xdevs.com/article/keithley-instruments/ does have neccessary strings for secret stuff:
00085BA0 SECR 00085BC0 ET CAL MENU yBB-RAM 5156 INIT 00085BE0 IALIZE OPTIONS NONE 6521 65 00085C00 22 5156 OPTION INSTALLED y 00085C20 HP-SHOOTER yCREDIT ENGLISH FRE 00085C40 NCH GERMAN yHIDDEN MESSAGES 00085C60 BKGD-AVG DISP-BURN-IN DQ-CAP y 00085C80 MSGS OPTION INIT-OPT CAL-STUFF 00085CA0 ySECRET MENU
And actual menu tree:
| OPTION | Override installed addon card detection | Force installed option (None / 6521 / 6522 / 5156) | |
| INIT-OPTS | Initialize EEPROMs for option cards | Initialize NVRAM or 5156 calibration option ROM | |
| CAL-STUFF | Various calibration items | Allow manually view/change cal data, calculate cal data, reset | |
| BKGD-AVG | Background averaging function | Enabled | Some digital filter, perhaps? Can be either ON or OFF |
| DISP-BURN-IN | Burn-in all VFD pixels | ||
| DQ_CAP | Use DQ capacitor? | Can be enabled or disabled |
| CREDIT | Shows The Design Team: credits | Lists all team members who work on Model 6517. ~30 seconds |
| HP-SHOOTER | Easter egg: 6517 shoots 4339 digits | 6517 shooting rockets to 4339, killing one digit every time :) |
| BB-RAM | Initialize BB-RAM (NVRAM) | Prepares DALLAS NVRAM to enable MEMx option installation |
| 5156 | Initialize EEPROM on 5156 | Initialize 5156 calibration module EEPROM |
| CAL-STUFF | Detail | Note |
| CONSTANTS | View calibration constants | |
| PROCEDURES | Calibration procedures? | |
| CALC&SAVE | Calculate calibration date and save | |
| FACTORY | Reset data to factory state | |
| RESTORE-ALL-DEFAULTS | Revert back to default data |
5156 Option is a calibration box with precision characterized resistances and capacitor.
Keithley Model 6220/6221 AC/DC source
“PULSE”+“RECALL” enables BIST and secret menu options for these instruments.
Keithley Model 6517B Electrometer
| ANALOG-REV | Write digital board revision ID into ROM | Stores digital board revision in calibration ROM | |
| DIGITAL-REV | Write analog board revision ID into ROM | Stores digital board revision in calibration ROM |
Keithley Model 7001/7002 switching frame
Switching frame Model 7001 have just credits information display.
To enter this mode do next
1. Ensure unit is powered off. 2. Press and hold button "6" and button "STORE" and power on unit 3. Enjoy credits information with developer team names listed 4. After showing all names, unit enters normal operation mode automatically
Model 7002 also have shooting game using switch LEDs and light pen, but since we don’t have this unit cannot give exact details on it.
Entrance to hidden mode should be same, hold buttons “6” and “STORE” on power on.
Keithley 2010/2182/2182A service mode
There is also some additional hidden modes available in Keithley instruments with single-line VFD screens.
On Keithley 2010 manufacturing calibration mode is activated by holding OPEN button while powering meter up. When meter booted in this mode full self-test diagnostics and debug menu are now available under SHIFT-TEST key combination. This is very handy for repairing old 2010 as sanity check of meters operation.
That’s if for now, if anyone have more details regarding Keithley gear, feel free to share via email (team[doggy]xdevs.com) or
on EEVBlog forums here
Modified: May 16, 2025, 3:39 p.m.
References
- EEVBlog forum : K1801 nanovolt preamp thread
- EEVBlog : Restoration glory of Keithley 2001 DMM
- Memory options on Keithley 2001/2002 series DMMs
- AVI: Keithley 2001 A-fw secret menu demo
- AVI: Keithley 2001 B-fw secret menu demo
- AVI: Keithley 2510 secret menu demo
- AVI: Keithley 2002 secret menu demo
- AVI: Keithley 7001 secret menu demo
- AVI: Keithley 2400 secret menu demo
- Documentation and firmwares
- EEVBlog forum : 6517 secret menu key entrance