Fluke 742A resistance standards repairs, stability and performance review


Intro

Every DMM and power measurement device are using electrical resistance standards to provide accurate and useful measurements. But if one need to verify precise instruments, highly stable and characterized fixed point resistance standards are required. Best examples of such artifacts are famous primary grade ESI SR104 or L&N 4210 units. These provide fixed resistance output, and essentially are custom made precision wirewound resistors mounted in nice rugged enclosure and provided with low-thermal binding posts for quality connections.

But in this article we will look at more common compact Fluke 742A resistance standards, designed for daily laboratory use. They might not be the best stability in the world and not the most stable devices, but way more practical and less demanding to typical handling. Fluke 742A performance still well suited for calibration of high-end long-scale multimeters, current sources, calibrators, thermometers and power measurement systems. To my knowledge this article is the most complete open public database of data and various measurements of Fluke 742A standard artifacts, with more than 25 units tested and some even repaired.

  • Good temperature stability, reducing need of airbath temperature control
  • Metrology grade with +18 °C to +28 °C temperature rating
  • Best 1 year specification is ± 4 µΩ/Ω for 10 and 19 kΩ variants
  • Each unit labeled with temperature α and β values
  • Compact size 86 mm x 105 mm x 127 mm and lightweight
  • 5-way low thermal binding posts, same as Fluke 57** calibrators or 8½-digit DMMs

In this article collection of available information and benchmarks for the available Fluke 742A units will be performed. This would represent performance those particular units, as example what can be expected from such a device, given typical laboratory conditions.

Index of tested Fluke 742A standards

Table below shows models that we measured directly and their test results summary. Models for 1 Ω and 10 kΩ are especially popular on secondary-market because they fit for artifact calibration needs of Fluke 5700A, 5720A and 5730A calibrators and Keysight 3458A multimeters.

Unit S/N Results Last known value Long-term drift (measured) Calibration Date, method
742A-1 Ω 5175026 Details 1.00005450 Ω ±0.7 µΩ/Ω +2.15 µΩ/Ω/year 9.MAR.2025 by xDevs, MI 6010B DCC
742A-1 Ω 5825005 Details 1.00003977 Ω ±0.4 µΩ/Ω +1.22 µΩ/Ω/year 7.AUG.2026 by xDevs, MI 6010B DCC
742A-1 Ω 6270023 Details 1.00004374 Ω ±0.43 µΩ/Ω TBD µΩ/Ω/year IET Labs external calibration, 27.OCT.2020
742A-1 Ω 6270023 Details 1.00004464 Ω ±0.33 µΩ/Ω TBD µΩ/Ω/year Internal xDevs.com DCC calibration, 20.MAR.2022
742A-1 Ω 7681001 Details 1.00000861 Ω ±0.4 µΩ/Ω TBD µΩ/Ω/year Internal xDevs.com calibration
742A-1 Ω 7791006 Details 1.000008026 Ω ±0.2 µΩ/Ω +0.358 µΩ/Ω/year 4.JAN.2026 by xDevs, MI 6010B DCC
742A-1.9 Ω 8272002 Details 1.90001979 Ω ±0.36 µΩ/Ω TBD µΩ/Ω/year 21.OCT.2021 by xDevs, G6675A DCC
742A-10 Ω 6330003 Details 10.0010222 Ω ±0.6 µΩ/Ω +2.20 µΩ/Ω/year 7.MAR.2025 by xDevs, MI 6010B DCC
742A-10 Ω 7170003 Details 10.0001170 Ω ±0.35 µΩ/Ω TBD µΩ/Ω/year Internal xDevs.com DCC calibration, 16.MAR.2022
742A-100 Ω 5445015 Details 100.001290 Ω ±0.5 µΩ/Ω +0.15 µΩ/Ω/year 6.MAR.2025 by xDevs, MI 6010B DCC
742A-1 kΩ 5505003 Details 1000.06676 Ω ±0.4 µΩ/Ω +1.35 µΩ/Ω/year 4.MAR.2025 by xDevs, MI 6010B DCC
742A-10 kΩ 5095049 Details 10000.0544 Ω ±0.3 µΩ/Ω +0.1 µΩ/Ω/year 4.MAR.2025 by xDevs, MI 6000A BVD
742A-10 kΩ 5065046 Details 10000.1428 Ω ±0.2 µΩ/Ω +0.46 µΩ/Ω/year 12.APR.2023 by xDevs, MI 6000A BVD
742A-10 kΩ 5190011 Details 10000.0916 Ω ±0.2 µΩ/Ω +0.3 µΩ/Ω/year 23.MAY.2024 by xDevs, corrected to MI data
742A-10 kΩ 5820007 Details 10000.09607 Ω ±0.3 µΩ/Ω +0.27 µΩ/Ω/year 7.AUG.2026 by xDevs, MI 6000A BVD
742A-10 kΩ 6250018 Details 10000.09331 Ω ±0.2 µΩ/Ω +0.264 µΩ/Ω/year 28.MAR.2026 by xDevs, MI 6000A BVD 316 µA
742A-10 kΩ 7633009 Details 10000.0381 Ω ±0.2 µΩ/Ω +0.153 µΩ/Ω/year 28.NOV.2025 by xDevs, MI 6000A BVD
742A-10 kΩ 7655004 Details 10000.1476 Ω ±0.4 µΩ/Ω +0.5 µΩ/Ω/year 17.APR.2025 by xDevs, MI 6000A BVD
742A-10 kΩ 7815007 Details 10000.0580 Ω ±0.2 µΩ/Ω +0.229 µΩ/Ω/year 3.APR.2026 by xDevs, MI 6000A BVD
742A-19 kΩ 6935001 Details 19000.1051 Ω ±0.3 µΩ/Ω +0.1 µΩ/Ω/year 12.APR.2023 by xDevs, MI 6000A BVD
742A-100 kΩ 7810004 Details 100.000995 kΩ ±0.4 µΩ/Ω +0.16 µΩ/Ω/year 12.APR.2023 by xDevs, MI 6000A BVD
742A-1 MΩ 5800002 Details 1.00000583 MΩ ±0.5 µΩ/Ω -0.33 µΩ/Ω/year 12.APR.2023 by xDevs, MI 6000A BVD
742A-10 MΩ 5095049 Details 10.0001009 MΩ ±0.8 µΩ/Ω -0.4 µΩ/Ω/year 12.APR.2023 by xDevs, MI 6000A BVD
742A-19 MΩ 6004003 Details 19.0001983 MΩ ±1.8 µΩ/Ω +1.16 µΩ/Ω/year 12.APR.2023 by xDevs, MI 6000A BVD
742A-19 MΩ 9103002 Details 19.0001510 MΩ ±1.2 µΩ/Ω TBD µΩ/Ω/year 18.JUL.2022 by xDevs, MI 6000A

Table 1: List of all measured standards in this article

Some additional custom values available from Fluke by special order. Fluke 742A standards are available for MSRP around ~$3.7 kUSD per piece (2018 pricing). Because of the support by xDevs members and our project friends access to their units for this review of multiple 742A was made possible.


Image 1: Array of Fluke 742A in natural habitat in air-bath, next to other commercial standards

There is also Fluke case designed to hold two 742A standards, Model 742A-7002. Case is available for $1315 USD.


Image 1: Array of Fluke 742A secondary resistance standards

Standard calibration fee for 742A with 17025 accreditation report from Fluke Everett calibration laboratory is $533 USD (2018). First report is included in price of the unit, if its ordered from Fluke new.

Many of these standards were calibrated and tested during xDevs.com annual Cal-Fest meetups in USA events, using high-resolution resistance bridge from Measurements International and Guildline.

Fluke 742A-1, 1 Ω model, S/N 6270023 from Fluke PSL, 02-NOV-1999

Unit is manufactured around end of 1999, if we believe label on the back. Since it is low ohmic resistance model, there are kelvin 4-wire port on the front and grounding post.

Fluke 742A-1 6270023 Value Unit
Nominal value 1 Ω
Operation temperature range +18 … +28 °C
Maximum voltage 0.5 V
Maximum current 0.5 A


Image 2: Front of the 742A-1 unit


Image 3-4: Back of the 742A-1 and top side view

This unit usually lives in plastic hardcase, such as Apache 2800. This case can house two of such standards with nice amount of padding on each side. This is useful to protect resistance standard from shipping stress and reduce risk of damage.


Image 5-6: 742A in safety of hard plastic case

For temperature coefficient verification, next setup is used:

Temperature stability test of this unit

After initial validation and testing this 742A-1 was shipped to external calibration lab for formal calibration on their DCC bridge setup.

It came back with assigned value 1.00004374 Ω measured on October 27, 2020 with uncertainty ±0.46 µΩ/Ω by IET Labs.

This resistor was recalibrated against freshly calibrated Ohm Labs 200 standard on 20.MAR.2022, with new assigned value 1.00004464 Ω with uncertainty ±0.33 µΩ/Ω. Measurement was performed at 100 mA current using MIL 6010B DCC bridge. Both reference standard and test 742A-1 were fixed at +23.00 °C temperature in air bath chamber.

Fluke 742A-10, 10 Ω model, S/N 7170003, received damaged

Fluke 742A-10 7170003 Value Unit
Nominal value 10 Ω
Operation temperature range +18 … +28 °C
Maximum voltage 1 V
Maximum current 0.1 A

This resistor unit was acquired damaged and here we have rare chance to look inside during and see repair steps. All four low thermal binding posts were replaced with new, but bases were cleaned up and reused. No alterations to resistor element were performed.

10 Ohm standard useful to verify current sources on 10mA range and DMMs which often have lowest range 10 Ω.

Damage on this unit was severe enough that current LO black 5-way low-thermal binding post was bent and half of the plastic cap missing. There is no alternative here but to replace whole post.

Label with factory calibrated value is gone, but warranty cal-sticker is intact :).

Temperature coefficient is not the best in the world, but pretty good for low-resistance standard.

Another look on damages to post. Time to disassemble the standard to gain internal access for posts replacements.

Fluke chassis is bit tricky and required careful peeling off front face mylar and “Fluke” logo label. There are four anodized black allen socket (3/32” imperial size) bolts holding front panel to the extruded case.

After getting bolts loose whole resistor block just slides out. To prevent mylar sticking back to the front panel small pieces of label paper was inserted on corners. Heating up corners to +45…50 °C helps softening adhesive that holds mylar.

Now we can see pretty hermetic wirewound resistors mounted between two tinned copper plates with predrilled holes. This 10 Ω standard built using next resistors:

  • 5 pcs Fluke 809236 100.1 ±0.02% +1.2 µΩ/Ω/°C TC resistors
  • 5 pcs Fluke 809235 100.1 ±0.02% with opposite -1.3 µΩ/Ω/°C TC resistors
  • 1 pcs Fluke 9989 Ω ±0.02% small resistor to provide fine-tune value

By simple math 10 resistors with 100.1 Ω nominal connected in parallel provide primary resistance 10.01 Ω. Using half resistor with positive temperature coefficient and other half with negative greatly reduce overall temperature coefficient of the assembly. To further improve accuracy small 9989 Ω nominal trim resistor connected in parallel to main resistor group drops total resistance to 9.9999790 Ω. Wide and thick copper bars may account for some microohms to get very close to ideal desired 10.000000 Ω.

Binding posts are same ones used in Fluke 5725A and similar Fluke calibration instruments. Rods with rotating knob are available to purchase as a spare part and also same to ones used in Fluke 5700A series calibrators. We had some spare posts from 5700A repair projects, so will replace here all four posts with brand new parts.

Bases with copper insert cleaned up nicely and installed back in unit.

It was interesting to note some sort of adhesive/glue applied on all nuts, perhaps some kind of loctite. It made disassembly somewhat tricky and required use of the torque wrench to get nuts loose.

Wires are soldered close to middle point of both copper bus bars, one pair of wires each side current and sense terminals. Wires were terminated with small ring lug.

Replacing connectors was easy after this step and here are few photos of happy Fluke 742A-10 resistor with some friends.

To obtain preliminary value of this resistance standard after repair and new post installation next equipment was used:

  • Keysight 34420A, calibrated 13.SEP.2020 against xDevs standards and Fluke 5720A/03
  • Fluke 5720A as high-stability +10.00005 and -10.00005 mA DC current source
  • Fluke 742A-1 reference resistor, calibrated as 1.00004374 Ω on October 27, 2020 with uncertainty ±0.46 µΩ/Ω by IET Labs.

Reference resistor was used to adjust Fluke 5720A output current to precisely 10.00000 mA in both polarity.

After doing few reversals of 10mA current initial value was obtained:

Better measurement was performed during comparison with calibrated Ohm Labs 200 1 Ω standard.

  • Ohm Labs 200 0.9999985 Ω , calibrated 1.MAR.2022 by Ohm Labs with U = ±0.3 µΩ/Ω, using MIL 6242D bridge and their Ohm Labs 200 1 Ω standard.
  • Comparisons at xDevs performed by MIL 6010B DCC bridge at 10mA (F742A-10) and 100mA (1 Ω standard). Reversal time was set at 12 seconds.
  • Both standards were placed in two separate active air baths .

Measurements of TCR with a two slow 3-day sweeps, one from +19 to +27 °C and second from +23 to +27 °C. Current result for α = +0.086 µΩ/Ω/°C and β = -0.030 µΩ/Ω/°C2. This is pretty close to factory sticker from 25 years ago.

Calibrated value at +23.0 °C on 16.MAR.2022 equals 10.0001170 Ω ±0.35 µΩ/Ω, with test current 10.0 mA.

Both 742A-1 and 742A-10 kit boxed together in transport case:

Live temperature corrected standard resistance calculator

The temperature correction chart in the lid of the each 742A unit is helpful to correct the resistance of the standard resistor for different ambient temperature effects. We added interactive real-time calculator for this article to aid with the calculation. Just fill in blue boxes from your 742A label information and enjoy the calculated value of expected resistance in green box. Formula used for 1 Ω example shown below:

RSTD = 1 Ω + (( (α × ΔTEMP) + (β × ΔTEMP2) + RSTD_DEV) × 1 Ω / 1-6) Ω

This realtime calculator accepts resistor parameters from label or calibration certificate to provide temperature-corrected output resistance at arbitrary environment temperature. Just enter α, β, deviation from nominal and desired actual temperature:

Nominal resistance +23 °C Ω
Alpha α +23 °C µΩ/Ω/°C
Beta β +23 °C µΩ/Ω/(°C)2
RSTD deviation +23 °C µΩ/Ω
Temperature to recalculate at °C

Standard Resistance output : Ω, µΩ/Ω from ideal Ω

This resistance value may be used as given at +23°C, if the change in resistance for the temperature range to be encountered is acceptable. For example, temperature variations less than ±2 °C from a nominal +23 °C would result in a worst case resistance deviation less than -0.3 µΩ/Ω. If this is an acceptable, then no temperature correction is required.

Here is also Python application to calculate resistance relations if you prefer to play with numbers more, based off ESI SR104 article .

Fluke 742A-1, 1 Ω model, S/N 7681001

Unit date of manufacture is unknown since no original factory sticker is present. Since it is low ohmic resistance model, there are kelvin 4-wire port on the front and grounding post.

Fluke 742A-1 7681001 Value Unit
Nominal value 1 Ω
Operation temperature range +18 … +28 °C
Maximum voltage 0.5 V
Maximum current 0.5 A


Image : Photos of the this 742A-1 unit

According to previous lab calibration sticker value of this standard was 1.000000 Ω measured with uncertainty ±4.9 µΩ/Ω at +23.0 °C with 100mA test current. This calibration was done on 2003/10/24 if we take typical 1 year calibration period (sticker only has due date and cryptic date calendar label from August 2001).

This resistor was recalibrated against freshly calibrated Ohm Labs 200 standard on 1.APR.2022, with new assigned value 1.00000861 Ω with uncertainty ±0.4 µΩ/Ω. Measurement was performed at 100 mA current using MIL 6010B DCC bridge. Both reference standard and test 742A-1 were fixed at +23.00 °C temperature in air bath chamber after initial warmup around 15 hours.

Fluke 742A-1 Ω model, S/N 7790003

Tom from a friendly calibration laboratory not far from xDevs HQ brought up his Fluke 742A 1 Ω for some xDevs ppm testing. There is no past history known about this particular standard, so calibration performed here would be the initial point in this xDevs database about 742As.

Standard has Fluke seal stickers on the sides. Overall resistor is in a pretty good shape externally.

Sadly on the back the original factory calibrated value was removed at some point during the life of this standard. This means we don’t know original factory calibrated value.

On the back we see quite interesting inventory label from Fluke with second number ST3305227 in additional to standard production serial number placement. I never seen such numbers on 742 before. TCR for this standard from factory are labelled as α -0.044 µΩ/Ω/°C, β -0.011 µΩ/Ω/°C^2. After sweeping temperature with MIL 6010B I’ve got following result:

Graphs are generated by xDevs.com TCkit application.

Model and serial Calibration date Assigned value Uncertainty Calculated drift Age
742A-1 7790003 17.APR.2025 1.00001165 Ω ± 0.8 µΩ/Ω Ref Ref

Fluke 742A-10 kΩ model, S/N 7655004

Tom from a friendly calibration laboratory near xDevs HQ brought up his Fluke 742A 10000 Ω for some xDevs ppm testing. There is no past history known about this particular standard, so calibration performed here would be the initial point in this xDevs database about 742As.

Standard has Fluke seal stickers on the sides.

Sadly on the back the original factory calibrated value was removed at some point during the life of this standard. This means we don’t know original factory calibrated value.

On the back we see quite interesting inventory label from Fluke with second number ST3305228 in additional to standard production serial number placement. I never seen such numbers on 742 before. TCR for this standard from factory are labelled as α +0.110 µΩ/Ω/°C, β -0.016 µΩ/Ω/°C^2.

Model and serial Calibration date Assigned value Uncertainty Calculated drift Age
742A-10k 7655004 5.SEP.2023 10000.13969 Ω ± ? µΩ/Ω Ref Ref
742A-10k 7655004 10.APR.2025 10000.14772 Ω ± 0.4 µΩ/Ω +0.80 µΩ/Ω, +0.5 µΩ/Ω/a 583 days
742A-10k 7655004 12.APR.2025 10000.14774 Ω ± 0.4 µΩ/Ω +0.8 µΩ/Ω 585 days
742A-10k 7655004 13.APR.2025 3:51:12 AM 10000.14782 Ω ± 0.4 µΩ/Ω +0.8 µΩ/Ω 586 days
742A-10k 7655004 13.APR.2025 3:51:19 AM 10000.14771 Ω ± 0.4 µΩ/Ω +0.8 µΩ/Ω 586 days
742A-10k 7655004 14.APR.2025 12:54:28 AM 10000.14767 Ω ± 0.4 µΩ/Ω +0.8 µΩ/Ω 587 days
742A-10k 7655004 14.APR.2025 3:38:56 PM 10000.14765 Ω ± 0.4 µΩ/Ω +0.8 µΩ/Ω 587 days
742A-10k 7655004 16.APR.2025 9:37:54 AM 10000.14783 Ω ± 0.4 µΩ/Ω +0.8 µΩ/Ω 589 days
742A-10k 7655004 17.APR.2025 12:57:20 PM 10000.14754 Ω ± 0.4 µΩ/Ω +0.8 µΩ/Ω 590 days
742A-10k 7655004 17.APR.2025 12:57:24 PM 10000.14775 Ω ± 0.4 µΩ/Ω +0.8 µΩ/Ω 590 days
742A-10k 7655004 17.APR.2025 12:57:28 PM 10000.14759 Ω ± 0.4 µΩ/Ω +0.8 µΩ/Ω 590 days
742A-10k 7655004 17.APR.2025 5:51:45 PM 10000.14788 Ω ± 0.4 µΩ/Ω +0.8 µΩ/Ω 590 days

Fluke 742A-10k, 10000 Ω model, S/N 5820007

Fluke 742A-10k 5820007 Value Unit
Nominal value 10000 Ω
Operation temperature range +18 … +28 °C
Maximum voltage 30 V
Maximum current 0.003 A

This resistance standard was acquired with heavy front damage on binding posts. The resistance measured OK, hopefully it is just the mechanical issue. It is relatively easy to fix as the binding posts are available as replacement parts.

Unit seen better days with completely destroyed sense terminals due to excessive torque applied to posts. Hopefully damage is only external and resistance elements still in good shape inside.

Disassembly of 742A is bit tricky due to hidden bolts that hold front and rear panels to square body metal chassis.

This 10 kΩ unit built using four hermetic Fluke wirewound resistance elements, P/N 806224. Each of these have equal nominal resistance 39992 Ω trimmed to ±50 µΩ/Ω precision.

Three elements are paralleled directly with copper shorting bars, while 4th connected with 31.759 Ω series mica wirewound resistor. With simple math we can conclude final resistance value should nominally equal 9999.9844 Ω which is pretty close to 10 kΩ nominal.

Selected elements are also not just random values, but actually matched for opposite TCR. There are +0.7 µΩ/Ω/°C, +0.2 µΩ/Ω/°C, -0.2 µΩ/Ω/°C and trimmed -0.7 µΩ/Ω/°C elements used here to give factory zero TCR output.

Overall repair was very similar to process shown above with Fluke 742A-10 box.

We will reuse original base inserts with copper bases and replace only binding post part with brand new parts.

Shortly after repair this standard was verified on APR.3.2022 for operation by direct measurement using Keysight 3458A DMM, which resulted for first value 10000.0700 Ω at nominal +23 °C with uncertainty ±3.6 µΩ/Ω. There was 6 hour soak time at +23.0 °C and +30 °C points to verify thermal delay behavior of resistor under test.

Standard with new binding posts was cleaned up, free now of any rogue labels and now ready for our performance benchmarks.

This resistor was recalibrated first against calibrated transfer standard ESI SR104 on 9.APR.2022, with new assigned value 10000.08649 Ω with uncertainty ±0.38 µΩ/Ω. Measurement was performed at 500 µA current using MIL 6000A resistance bridge with 12 seconds settling time per 15 readings and battery powered Fluke 732B 10V DC source. Detector in this measurement was again 8½-digit Keysight 3458A voltmeter.

Both reference standard and test 742A-10k were fixed at +23.00 °C temperature in air bath chamber after initial warmup around 24 hours.

More details about DIY air bath available in this dedicated article . Accurate temperature of Fluke 742A under test was obtained from Honeywell HEL-705 Pt1000 sensor placed directly in chassis 742As’ grounding post. Fluke 1529 Chub-E4 used to collect temperature data.

Temperature coefficient was also remeasured against same reference standard with MIL 6010B DCC bridge. For this test two air bath chambers were used, one at fixed +23.00 °C temperature with transfer Guildline 9330 1000.03470 Ω standard. Temperature in second air bath with DUT Fluke 742A was slowly changed from +17.7 °C to +30.3 °C over duration of 68 hours. Ramp up and ramp down temperatures were equal in duration, with ΔT speed 0.0062 °C/minute.

Resistance measurements were collected during temperature sweep, resulting in RAW data, shown in time-scale below. RAW-file available by clicking on plot.

After some basic math correlation of Fluke 742A-10k measured resistance to temperature was established. Temperature coefficient of this resistance standard now established at α = +0.006 µΩ/Ω/°C ; β = -0.024 µΩ/Ω/°C2 and zero TCR intersection point at +23.1 °C.

Manufacture date of this resistance standard is June 1993, which together with the calibration sticker from Canadian Hydro-Québec Research Institute with 2011 calibration allows us to give educated guesstimate of annual drift for this unit.

Daily DriftRSTD = Current value +8.04 µΩ/Ω / Time between current calibration April 2022 and manufacture date June 1993 (10600 days) = +8.04 / 10600 = +0.000758 µΩ/Ω/day

By running simple math we can estimate annual linear drift +0.000758 µΩ/Ω/day * 365 days = +0.277 µΩ/Ω/year. This is well inside ±8 µΩ/Ω/year specified in Fluke instruction manual for Model 742A standards. Also running same math from ITEQ calibration sticker dated June 3, 2011 we can establish even smaller figure : +1.84 µΩ/Ω / 3963 days = +0.000464 µΩ/Ω/day = +0.169 µΩ/Ω/year. All measurements at xDevs lab from 2022 and later are referenced to the golden lab standard, which is ESI SR104 primary standard which was calibrated by accredited Measurements International laboratory number of times in recent years with very low absolute uncertainty ±0.1 µΩ/Ω.

Model and serial Calibration date Assigned value Uncertainty Calculated drift Points in calculation Age
742A-10k 5820007 JUN.1993 10000.0060 Ω ± 1.00 µΩ/Ω Ref Ref
742A-10k 5820007 3.JUN.2011 10000.0680 Ω ± 0.75 µΩ/Ω +6.2 µΩ/Ω, +0.345 µΩ/Ω/a 6563 days
742A-10k 5820007 14.APR.2022 10000.0865 Ω ± 0.38 µΩ/Ω +8.05 µΩ/Ω, +0.279 µΩ/Ω/a 10531 days
742A-10k 5820007 9.MAY.2022 10000.0872 Ω ± 0.35 µΩ/Ω +8.12 µΩ/Ω, +0.281 µΩ/Ω/a 115 10556 days
742A-10k 5820007 8.MAR.2023 10000.0880 Ω ± 0.4 µΩ/Ω +8.20 µΩ/Ω, +0.276 µΩ/Ω/a 4000 10859 days
742A-10k 5820007 14.APR.2023 10000.0886 Ω ± 0.2 µΩ/Ω +8.26 µΩ/Ω, +0.276 µΩ/Ω/a 256 10910 days
742A-10k 5820007 7.AUG.2023 10000.0871 Ω ± 0.15 µΩ/Ω +8.11 µΩ/Ω, +0.268 µΩ/Ω/a 128 11024 days
742A-10k 5820007 9.JAN.2024 10000.0886 Ω ± 0.18 µΩ/Ω +8.26 µΩ/Ω, +0.27 µΩ/Ω/a 128 11179 days
742A-10k 5820007 23.MAY.2024 10000.0904 Ω ± 0.2 µΩ/Ω +8.44 µΩ/Ω, +0.272 µΩ/Ω/a 128 11314 days
742A-10k 5820007 22.JUL.2024 10000.0895 Ω ± 0.25 µΩ/Ω +8.35 µΩ/Ω, +0.268 µΩ/Ω/a 1676 11374 days
742A-10k 5820007 2.AUG.2024 10000.0894 Ω ± 0.25 µΩ/Ω +8.34 µΩ/Ω, +0.267 µΩ/Ω/a 128 11385 days
742A-10k 5820007 17.DEC.2024 10000.0901 Ω ± 0.25 µΩ/Ω +8.41 µΩ/Ω, +0.266 µΩ/Ω/a 200 11509 days
742A-10k 5820007 4.MAR.2025 10000.09033 Ω ± 0.3 µΩ/Ω +8.433 µΩ/Ω, +0.265 µΩ/Ω/a 256 11586 days
742A-10k 5820007 31.MAR.2025 10000.0902 Ω ± 0.3 µΩ/Ω +8.42 µΩ/Ω, +0.264 µΩ/Ω/a 256 11613 days
742A-10k 5820007 28.MAR.2026 10000.09484 Ω ± 0.2 µΩ/Ω +8.884 µΩ/Ω, +0.27 µΩ/Ω/a 256 11988 days
742A-10k 5820007 4.JUN.2026 10000.09456 Ω ± 0.2 µΩ/Ω +8.856 µΩ/Ω, +0.268 µΩ/Ω/a 256 12056 days
742A-10k 5820007 7.AUG.2026 10000.09607 Ω ± 0.3 µΩ/Ω +8.x µΩ/Ω, +0.27 µΩ/Ω/a 256 12056 days, after overheat

We plan to use this 10 kΩ as a travel standard between xDevs members for resistance calibrations and various studies. It is small, easy to ship and have decent performance to meet most of practical metrology demands. Average annual stability of this standard is estimated to be about quite decent +0.27 µΩ/Ω/year.

This same drift data can be presented in graphical chart form as well for more visual consumption:

Update with latest year 2024 data:

Overall, this is quite good and happy 742A-10k after the binding posts replacement, proving the stability and robust design of Fluke transfer resistance standards.

Fluke 742A-10 kΩ model, S/N 5190011

Fluke 742A-10k 5190011 Value Unit
Nominal value 10000 Ω
Operation temperature range +18 … +28 °C
Maximum voltage 30 V
Maximum current 0.003 A

This resistance standard is provided for review and testing by the friendly lab, codenamed here as “DN”.

This DUT Fluke standard was recently calibrated by Measurements International on JUN.26.2023 using their 6010D system and reference SR104. Excerpt from the calibration report is attached below for the reference information only:

Calibration was performed at higher 1mA current, so some self-heating effects might be included in the result. Typically at xDevs.com we calibrated 10 kΩ standards at lower 100 or 316 µA currents to closer simulate actual use with DMMs and calibrators. 316 µA is also maximum current for 10 kΩ to allow comparison at 10 mW power for 1 kΩ by using DCC bridge, such as MI 6010B.

This resistor was recalibrated again by xDevs.com here first against prototype MP-1kΩ transfer standard during 6.SEP.2023 to 10.SEP.2023, with new assigned value 10000.0882 Ω with expanded uncertainty ±0.3 µΩ/Ω. Measurement was performed at 316 µA current using MIL 6010B resistance bridge with 20 seconds settling time per 17119 readings. Temperature in air bath with DUT Fluke 742A was slowly changed from +17.8 °C to +28.2 °C over duration of 94 hours. Ramp up and ramp down temperatures were equal in duration, with ΔT speed 0.0057 °C/minute.

Both transfer standard and test 742A-10k were swept at in air bath chamber after initial warmup over 24 hours. Transfer standard MP-1k was freshly calibrated from Fluke SL935 reference standard to obtain current value. Transfer standard MP-1k does not exhibit temperature coefficient (measured α = -0.00002 µΩ/Ω/°C ; β = -0.00014 µΩ/Ω/°C2). More details about DIY air bath available in this dedicated article . Accurate temperature of Fluke 742A under test was obtained from Honeywell HEL-705 Pt1000 sensor placed directly in chassis 742As’ grounding post. Fluke 1529 Chub-E4 used to collect temperature data.

Resistance measurements were collected during temperature sweep, resulting in RAW data, shown in time-scale below. RAW-file with all data points is available by clicking on plot. Small +0.22 µΩ/Ω hysteresis was observed after the temperature sweep experiment which recovered to previous values about 20 hours after the run (not shown on chart).

After some basic math correlation of Fluke 742A-10k measured resistance to temperature change was established. Temperature coefficient of this resistance standard now established at α = +0.031 µΩ/Ω/°C ; β = -0.032 µΩ/Ω/°C2 and zero TCR intersection point at +23.5 °C. This is slightly different than old α = +0.000 µΩ/Ω/°C ; β = -0.028 µΩ/Ω/°C2 from Fluke’s factory certificate from 1991 year, but could be explained by different test current.

Standard’s measured value also rechecked with separate binary-voltage divider bridge system, built around Measurements International 6000A, 20V source and HP 3458A as detector. It is pretty impressive to see very good degree of agreement between our own in-house measurement and MI accredited calibration 79 days ago. Such agreement is easy to calculate as:

((RMI / RxDevs) – 1) * 100000 ; or ((10000.0888 Ω / 10000.0882 Ω) – 1)*1e6 = -0.06 µΩ/Ω

This result is even below our best measurement uncertainty. In-house MI 6000A system compared Fluke 742A-10k 5190011 to xDevs.com’s primary ESI SR104 four times, 32 samples each direction during 18 hours of measurements. 1mA test current was established during the transfers to match accredited calibartion parameters.

Standard deviation of each data set was well below 0.03 µΩ/Ω. Set of these measurements provided with value 10000.0900 Ω with expanded uncertainty ±0.2 µΩ/Ω. Both ESI SR104 and DUT resistor were stabilized in active air bath at +23.0 °C ±0.2 °C. This data was used to correct 6010B results with account for measured TCR and final value for Fluke 742A-10k 5190011 assigned as 10000.0904 Ω ±0.2 µΩ/Ω.

Daily DriftRSTD = Current value +9.6 µΩ/Ω / Time between current calibration September 2023 and manufacture date March 1991 (11859 days) = +9.6 / 11859 = +0.00081 µΩ/Ω/day

By running simple math we can estimate annual linear drift +0.00081 µΩ/Ω/day * 365 days = +0.295 µΩ/Ω/year. This is well inside ±5 µΩ/Ω/year specified in Fluke instruction manual for Model 742A-10k standards.

Model and serial Calibration date Assigned value Uncertainty Calculated change and drift Points in calculation Age
742A-10k 5190011 26.MAR.1991 9999.991 Ω ± 1 µΩ/Ω Ref Ref
742A-10k 5190011 07.OCT.1992 9999.97 Ω ± 4 µΩ/Ω -2.1 µΩ/Ω, -1.366 µΩ/Ω/a 561 days
742A-10k 5190011 26.JUN.2023 by MI, MI 6010D at 1mA 10000.0888 Ω ± 0.1 µΩ/Ω +9.78 µΩ/Ω, +0.303 µΩ/Ω/a 150 11780 days
742A-10k 5190011 6.SEP.2023 by xDevs, MI 6010B at 0.316 mA 10000.0882 Ω ± 0.2 µΩ/Ω +9.72 µΩ/Ω, +0.3 µΩ/Ω/a 17119 11852 days
742A-10k 5190011 13.SEP.2023 by xDevs, MI 6000A at 1mA 10000.0867 Ω ± 0.15 µΩ/Ω +9.6 µΩ/Ω, +0.295 µΩ/Ω/a 125 11859 days
742A-10k 5190011 25.DEC.2023 by xDevs, MI 6010B at 0.32 mA 10000.0889 Ω ± 0.2 µΩ/Ω +9.79 µΩ/Ω, +0.298 µΩ/Ω/a 20000 11962 days
742A-10k 5190011 27.MAR.2024 by xDevs, MI 6000A at 0.5 mA 10000.0901 Ω ± 0.2 µΩ/Ω +9.91 µΩ/Ω, +0.300 µΩ/Ω/a 64 12055 days
742A-10k 5190011 28.MAR.2024 by xDevs, MI 6000A at 0.5 mA 10000.0900 Ω ± 0.2 µΩ/Ω +9.90 µΩ/Ω, +0.300 µΩ/Ω/a 64 12056 days
742A-10k 5190011 17.MAY.2024 by xDevs, MI 6000A at 0.316 mA 10000.0904 Ω ± 0.2 µΩ/Ω +9.94 µΩ/Ω, +0.300 µΩ/Ω/a 64 12106 days
742A-10k 5190011 21.MAY.2024 by xDevs, MI 6000A at 0.500 mA 10000.0906 Ω ± 0.2 µΩ/Ω +9.96 µΩ/Ω, +0.300 µΩ/Ω/a 64 12110 days
742A-10k 5190011 23.MAY.2024 by xDevs, MI 6000A at 1.000 mA 10000.0907 Ω ± 0.2 µΩ/Ω +9.97 µΩ/Ω, +0.300 µΩ/Ω/a 50 12112 days
742A-10k 5190011 23.MAY.2024 by xDevs, MI 6000A at 0.316 mA 10000.0916 Ω ± 0.2 µΩ/Ω +10.06 µΩ/Ω, +0.303 µΩ/Ω/a corrected with drift 12112 days

Agreement within -0.15 µΩ/Ω between different measurement methods using MI 6010B and MI 6000A was found acceptable, considering different test currents and temperatures. Average annual stability of this standard is estimated to be about quite decent +0.3 µΩ/Ω/year, which is quite similar to our own 742A-10k tested before. This same drift data can be presented in graphical chart form as well for more visual consumption:

Difference of the output value from using different test current was also investigated for this 742A-10k S/N 5190011 unit. This was achieved by using thermally stabilized transfer resistance standard as fixed point and changing current setting for Measurements International 6010B DCC bridge. Result is available on the chart below.

The deviation of the median value between 1000 µA and 316 µA was just 33 nΩ/Ω which is essentially invisible for most of the resistance calibration labs who use digital multimeters. Even high-end 8½-digit multimeter such as Keysight 3458A or Fluke 8588A has magnitude larger noise level for resistance function.

Fluke 742A-1, 1 Ω model, S/N 5825005

Fluke 742A-1 5825005 Value Unit
Nominal value 1 Ω
Operation temperature range +18 … +28 °C
Maximum voltage 0.5 V
Maximum current 0.5 A

This resistor was received and tested first against freshly calibrated Ohm Labs 200 standard on 2.APR.2022, with new assigned value 1.00003507 Ω ([1] , [2] ) with uncertainty ±0.4 µΩ/Ω. Measurement was performed at 100 mA current using MIL 6010B DCC bridge. Both reference standard and test 742A-1 were fixed at +23.00 °C temperature after initial warmup around 24hours.

Same setup as in Fluke 742-10k case used to remeasure TCR as well. Two air bath chambers were used, one at fixed +23.00 °C temperature with reference Ohm Labs 200 0.9999985 Ω standard. Omega RTDCAP 3-wire 100 Ω temperature sensor was placed in Ohm Labs 200 thermometer well for accurate temperature data for correlation math.

Temperature in second air bath with DUT Fluke 742A was slowly changed from +17.5 °C to +30.3 °C over duration of 118 hours. Ramp up and ramp down temperatures were equal in duration, with ΔT speed 0.0056 °C/minute. Fluke 1529 Chub-E4 used to collect temperature data.

Resistance measurements were collected during temperature sweep, resulting in RAW data, shown in time-scale below. RAW-file available by clicking on plot.

After some basic math correlation of 742A-1 measured resistance to temperature was established. Temperature coefficient of this resistance standard now established at α = -0.109 µΩ/Ω/°C ; β = -0.040 µΩ/Ω/°C2 and zero TCR intersection point at +21.6 °C.

Estimated manufacture date of this resistor is June 1993, which together with the calibration sticker from Canadian Hydro-Québec Research Institute with 2011 calibration allows us to give educated guesstimate of annual drift for this particular Fluke 742A-1 standard.

Daily DriftRSTD = Current value +35.07 µΩ/Ω / Time between current calibration April 2022 and manufacture date June 1993 (10600 days) = +35.07 / 10600 = +0.00331 µΩ/Ω/day

By running simple math we can estimate annual linear drift +0.00331 µΩ/Ω/day * 365 days = +1.20 µΩ/Ω/year. This is well inside ±8 µΩ/Ω/year specified in Fluke instruction manual for Model 742A standards. Also running same math from ITEQ calibration sticker dated June 3, 2011 we can establish similar figures : +12.37 µΩ/Ω / 3963 days = +0.00312 µΩ/Ω/day = +1.14 µΩ/Ω/year. This gives some additional confidence that resistor survived these years well enough to keep that stability to present date.

Now this resistance standard is ready to keep precious 1 Ω and current µΩ/Ωs for xDevs.com lab, together with ovenized Fluke SL935.

This unit was recalibrated during xDevs.com RepairFest 2023 against freshly calibrated Fluke SL935 ovenized standard on 12.APR.2023, with new assigned value 1.00003626 Ω with uncertainty ±0.1 µΩ/Ω.

Measurement was performed at 100 mA current using MIL 6010B DCC bridge. Test 742A-1 was sweeped from +18.0 °C to +28.0 °C temperature after initial warmup around >24hours. This calibration point provided with the estimated drift of this 742A-1 at +1.2 µΩ/Ω/year, which is in line previous linear prediction perfectly. Temperature coefficient was also verified as result, with new α = -0.110 µΩ/Ω/°C ; β = -0.040 µΩ/Ω/°C2* and zero TCR intersection point at +22.0 °C.

Model and serial Calibration date Assigned value Uncertainty Calculated drift Points in calculation Age
742A-1 5825005 JUN.1993 1.0000002 Ω, 100mA ± 2.0 µΩ/Ω Ref Ref
742A-1 5825005 3.JUN.2011 1.0000227 Ω, 100mA ± 0.7 µΩ/Ω +22.5 µΩ/Ω, +1.14 µΩ/Ω/a 6563 days
742A-1 5825005 2.APR.2022 1.00003507 Ω, 100mA ± 0.35 µΩ/Ω +34.9 µΩ/Ω, +1.2 µΩ/Ω/a 115 10532 days
742A-1 5825005 12.APR.2023 1.00003626 Ω, 100mA ± 0.1 µΩ/Ω +36.1 µΩ/Ω, +1.2 µΩ/Ω/a 12000 10907 days
742A-1 5825005 7.AUG.2023 1.00003702 Ω, 100mA ± 0.25 µΩ/Ω +36.8 µΩ/Ω, +1.22 µΩ/Ω/a 8700 11024 days
742A-1 5825005 23.JAN.2024 1.00003758 Ω, 100mA expected ± 0.3 µΩ/Ω +37.4 µΩ/Ω, +1.22 µΩ/Ω/a ? 11193
742A-1 5825005 6.JAN.2025 1.00003916 Ω, 100mA expected ± 0.3 µΩ/Ω +38.96 µΩ/Ω, TBD µΩ/Ω/a ? 11542
742A-1 5825005 15.JAN.2025 1.00003820 Ω, 31.6 mA ± 0.4 µΩ/Ω +38.0 µΩ/Ω, +1.20 µΩ/Ω/a 2242 11538 days
742A-1 5825005 9.MAR.2025 1.00003799 Ω, 100 mA ± 0.6 µΩ/Ω +37.8 µΩ/Ω, +1.19 µΩ/Ω/a 11961 11591 days
742A-1 5825005 4.JAN.2025 1.00003868 Ω, 50 mA ± 0.4 µΩ/Ω +xx.x µΩ/Ω, +x.xx µΩ/Ω/a xxxxx xxxxx days
742A-1 5825005 7.AUG.2025 1.00003977 Ω, 50 mA ± 0.4 µΩ/Ω +xx.x µΩ/Ω, +x.xx µΩ/Ω/a xxxxx xxxxx days, after overheat

This same drift data can be presented in graphical chart form as well for more visual consumption:

Fluke 742A-1 Ω model, S/N 5340003

Fluke 742A-1 Value Unit
Nominal value 1 Ω
Operation temperature range +18 … +28 °C
Maximum voltage 0.5 V
Maximum current 500 mA

This unit belongs to xDevs-FL and was acquired from secondary market. It started participation in CalFest intercomparisons from 2021 between xDevs HQ and FL with Guildline 6675 and MI 6010B DCC bridges.

Unit came with factory labels, so we can see original value was 1.0000006 Ω and TCR coefficients equal α = -0.060 µΩ/Ω/°C ; β = -0.032 µΩ/Ω/°C2.

Fluke 742A-1 Ω model, S/N 5175026

Fluke 742A-1 Value Unit
Nominal value 1 Ω
Operation temperature range +18 … +28 °C
Maximum voltage 0.5 V
Maximum current 500 mA

This unit belongs to xDevs-FL and was acquired from secondary market. It started participation in CalFest intercomparisons from 2021 between xDevs HQ and FL with Guildline 6675 and MI 6010B DCC bridges.

Unit came with factory labels, so we can see original value was 1.0000005 Ω and TCR coefficients equal α = 0.000 µΩ/Ω/°C ; β = -0.044 µΩ/Ω/°C2.

Model and serial Calibration date Assigned value Uncertainty Calculated drift Notes Age
742A-1 5175026 Unknown 1.0000005 Ω ± 0.3 µΩ/Ω ? Fluke ?
742A-1 5175026 12.APR.2023 by xDevs 1.00005044 Ω ± 0.3 µΩ/Ω +49.9 µΩ/Ω 6010B DCC vs Fluke SL935-1, 100mA 0 days
742A-1 5175026 9.MAR.2025 by xDevs 1.00005450 Ω ± 0.7 µΩ/Ω +54.0 µΩ/Ω, +2.147 µΩ/Ω/a 6010B DCC vs Fluke 742A-10 6330003, 100mA 697 days

Fluke 742A-1 Ω model, S/N 7791006

Fluke 742A-1 7791006 Value Unit
Nominal value 1 Ω
Operation temperature range +18 … +28 °C
Maximum voltage 0.5 V
Maximum current 0.5 A

According to labels this standard was manufactured and calibrated by Fluke lab on 9 May, 2001. We have additional calibrations sticker with values from from IRSST .

Looking at the back we can find good low temperature coefficient.

Model and serial Calibration date Assigned value Uncertainty Calculated drift Points in calculation Age
742A-1 7791006 9.MAY.2001 0.9999991 Ω ± 1.00 µΩ/Ω Ref Ref
742A-1 7791006 24.OCT.2004 0.999997 Ω ± 4.9 µΩ/Ω -2.1 µΩ/Ω, -0.606 µΩ/Ω/a 1264 days
742A-1 7791006 25.APR.2022 by Datron 1281 0.9999751 Ω ± 4 µΩ/Ω 7656 days
742A-1 7791006 13.JUL.2022 by xDevs.com 1.00000570 Ω ± 0.4 µΩ/Ω +6.6 µΩ/Ω, +0.311 µΩ/Ω/a 11605 by MI 6010B DCC 7735 days
742A-1 7791006 8.APR.2023 by xDevs.com 1.00000669 Ω ± 0.4 µΩ/Ω +7.59 µΩ/Ω, +0.346 µΩ/Ω/a 2879 by MI 6010B DCC, 100 mA 8004 days
742A-1 7791006 29.AUG.2024 by xDevs.com 1.00000631 Ω ± 0.4 µΩ/Ω +7.21 µΩ/Ω, +0.309 µΩ/Ω/a 1471 by MI 6010B DCC, 50 mA 8513 days
742A-1 7791006 4.JAN.2026 by xDevs.com 1.00000724 Ω ± 0.4 µΩ/Ω +8.14 µΩ/Ω, +0.330 µΩ/Ω/a 1111 by MI 6010B DCC, 50 mA 9006 days
742A-1 7791006 3.APR.2026 by xDevs.com 1.000008026 Ω ± 0.2 µΩ/Ω +8.926 µΩ/Ω, +0.358 µΩ/Ω/a 1240 by MI 6010 DCC, 50 mA 9095 days

All calibrations performed at +23.0 °C air-bath chamber. No TCR measurements performed at this time for this standard.

Fluke 742A-1.9 Ω model, S/N 8272002

Fluke 742A-1.9 Value Unit
Nominal value 1.9 Ω
Operation temperature range +18 … +28 °C
Maximum voltage 0.38 V
Maximum current 200 mA

This unit belongs to xDevs-FL and was acquired from secondary market. It started participation in CalFest intercomparisons from 2021 between xDevs HQ and FL with Guildline 6675 and MI 6010B DCC bridges.

Unit came with factory labels, so we can see original value was 1.8999780 Ω and TCR coefficients equal α = -0.015 µΩ/Ω/°C ; β = -0.022 µΩ/Ω/°C2. There are calibration labels from Verizon ERS for report number V0311964 from 31 July, 2007 but no actual calibration value or copy available.

Model and serial Calibration date Assigned value Uncertainty Calculated drift Notes Age
742A-1.9 8272002 Unknown 1.8999780 Ω ± 1.0 µΩ/Ω ? Fluke ?
742A-1.9 8272002 21.OCT.2021 by xDevs 1.90001979 Ω ± 0.36 µΩ/Ω µΩ/Ω, µΩ/Ω/a DCC vs NLab 742-1, 72.6mA days

Fluke 742A-10 Ω model, S/N 6330003

Fluke 742A-10 Value Unit
Nominal value 10 Ω
Operation temperature range +18 … +28 °C
Maximum voltage 1 V
Maximum current 100 mA

This unit belongs to xDevs-FL and was acquired from secondary market. It started participation in CalFest intercomparisons from 2021 between xDevs HQ and FL with Guildline 6675 and MI 6010B DCC bridges.

Unit came without any factory labels except serial number, so we don’t know original calibration values or TCR.

Model and serial Calibration date Assigned value Uncertainty Calculated drift Notes Age
742A-10 6330003 21.OCT.2021 by xDevs 10.00094775 Ω ± 0.28 µΩ/Ω Reference DCC vs NLab 742-1, 10mA 0 days
742A-10 6330003 15.AUG.2022 by xDevs 10.00094663 Ω ± 0.4 µΩ/Ω -0.112 µΩ/Ω, -0.137 µΩ/Ω/a 6010B DCC vs Fluke SL935-1, 10mA 298 days
742A-10 6330003 12.APR.2023 by xDevs 10.00097617 Ω ± 0.3 µΩ/Ω +2.842 µΩ/Ω, +1.928 µΩ/Ω/a 6010B DCC vs Fluke SL935-1, 10mA 538 days
742A-10 6330003 7.MAR.2025 by xDevs 10.0010222 Ω ± 0.6 µΩ/Ω +7.444 µΩ/Ω, +2.203 µΩ/Ω/a 6010B DCC vs Fluke 742A-100 5445015, 10mA 1233 days

Fluke 742A-100 Ω model, S/N 5445015

Fluke 742A-100 Value Unit
Nominal value 100 Ω
Operation temperature range +18 … +28 °C
Maximum voltage 2 V
Maximum current 20 mA

This unit belongs to xDevs-FL and was acquired from secondary market. It started participation in CalFest intercomparisons from 2021 between xDevs HQ and FL with Guildline 6675 and MI 6010B DCC bridges.

Unit came without value label but has TCR and serial number. TCR coefficients α = +0.007 µΩ/Ω/°C ; β = -0.041 µΩ/Ω/°C2. There are calibration labels from Measurement Assurance Technology company from 2 April, 2009 but no actual calibration value or copy available.

Interesting to see original Fluke seal stickers on the side of the standard.

Model and serial Calibration date Assigned value Uncertainty Calculated drift Notes Age
742A-100 5445015 21.OCT.2021 by xDevs 100.001240 Ω ± 0.4 µΩ/Ω Reference 10 hours, DCC vs NLab 4310AF-100 0 days
742A-100 5445015 15.AUG.2022 by xDevs 100.001196 Ω ± 0.4 µΩ/Ω -0.44 µΩ/Ω, -0.539 µΩ/Ω/a 6010B DCC vs Fluke 742A-1k, 10mA 298 days
742A-100 5445015 12.APR.2023 by xDevs 100.001220 Ω ± 0.4 µΩ/Ω -0.20 µΩ/Ω, -0.136 µΩ/Ω/a 6010B DCC vs Fluke 742A-1k, 10mA 538 days
742A-100 5445015 6.MAR.2025 by xDevs 100.001290 Ω ± 0.5 µΩ/Ω +0.50 µΩ/Ω, +0.148 µΩ/Ω/a 6010B DCC vs Fluke 742A-1k, 10mA 1232 days

Fluke 742A-1 kΩ model, S/N 5505003

Fluke 742A-1k Value Unit
Nominal value 1000 Ω
Operation temperature range +18 … +28 °C
Maximum voltage 10 V
Maximum current 10 mA

This unit belongs to xDevs-FL and was acquired from secondary market. It started participation in CalFest intercomparisons from 2021 between xDevs HQ and FL with Guildline 6675 and MI 6010B DCC bridges.

Unit came without value label but has TCR and serial number. TCR coefficients α = +0.050 µΩ/Ω/°C ; β = -0.018 µΩ/Ω/°C2.

Model and serial Calibration date Assigned value Uncertainty Calculated drift Notes Age
742A-1k 5505003 21.OCT.2021 by xDevs 1000.06219 Ω ± 0.35 µΩ/Ω Reference 4 hours, DCC vs NLab 4310AF-100 0 days
742A-1k 5505003 15.AUG.2022 by xDevs 1000.06247 Ω ± 0.5 µΩ/Ω 0.28 µΩ/Ω, +0.343 µΩ/Ω/a 6010B DCC vs Fluke SL935-10k, 3.16 mA 298 days
742A-1k 5505003 12.APR.2023 by xDevs 1000.06333 Ω ± 0.3 µΩ/Ω 1.14 µΩ/Ω, +0.773 µΩ/Ω/a 6010B DCC vs Fluke SL935-10k, 3.16 mA 538 days
742A-1k 5505003 4.MAR.2025 by xDevs 1000.06676 Ω ± 0.4 µΩ/Ω 4.57 µΩ/Ω, +1.356 µΩ/Ω/a 6010B DCC vs Fluke SL935-10k, 3.16 mA 1230 days

Fluke 742A-10 kΩ model, S/N 5095049

Fluke 742A-10k Value Unit
Nominal value 10000 Ω
Operation temperature range +18 … +28 °C
Maximum voltage 30 V
Maximum current 3 mA

This unit belongs to xDevs-FL and was acquired from secondary market. It started participation in CalFest intercomparisons from 2022 between xDevs HQ and FL with MI 6000A BVD bridge system.

Unit came without value label but has TCR and serial number. TCR coefficients α = +0.080 µΩ/Ω/°C ; β = -0.016 µΩ/Ω/°C2. It has also calibration sticker dated 26 March, 2009 from Fluke calibration lab. Also presence of SI-1990 sticker suggests that it was manufactured and certified at least after 1990 year.

Model and serial Calibration date Assigned value Uncertainty Calculated drift Notes Age
742A-10k 5095049 12.APR.2023 by xDevs 10000.0529 Ω ± 0.2 µΩ/Ω Reference 6000A BVD vs Fluke SL935-10k, 0.5 mA 0 days
742A-10k 5095049 4.MAR.2025 by xDevs BVD 10000.05436 Ω ± 0.3 µΩ/Ω +0.146 µΩ/Ω, +0.077 µΩ/Ω/a 6000A BVD vs Fluke SL935-10k, 0.5 mA 692 days

Fluke 742A-10 kΩ model, S/N 5065046

Fluke 742A-10k Value Unit
Nominal value 10000 Ω
Operation temperature range +18 … +28 °C
Maximum voltage 30 V
Maximum current 3 mA

This unit belongs to xDevs-FL and was acquired from secondary market. It started participation in CalFest intercomparisons from 2021 between xDevs HQ and FL with Guildline 6675 and MI 6010B DCC bridges.

Unit came with all good labels and has TCR with value. Calibration value was 9999.993 Ω with TCR coefficients α = +0.170 µΩ/Ω/°C ; β = -0.018 µΩ/Ω/°C2. It has also calibration sticker from Fluke Primary Standards Laboratory dated 1 October 1990.

Model and serial Calibration date Assigned value Uncertainty Calculated drift Notes Age
742A-10k 5065046 1.OCT.1990, factory 9999.993 Ω ± 1.0 µΩ/Ω Ref Original factory label Ref
742A-10k 5065046 21.OCT.2021 by xDevs 10000.1376 Ω ± 0.3 µΩ/Ω +14.46 µΩ/Ω, +0.465 µΩ/Ω/a 4 hours, DCC vs NLab 4310AF-100 11343 days
742A-10k 5065046 12.JUL.2022 by xDevs 10000.1397 Ω ± 0.25 µΩ/Ω +14.67 µΩ/Ω, +0.461 µΩ/Ω/a 6010B DCC vs Fluke SL935-10k, 0.316 mA 11607 days
742A-10k 5065046 12.APR.2023 by xDevs 10000.1428 Ω ± 0.2 µΩ/Ω +14.98 µΩ/Ω, +0.460 µΩ/Ω/a 6000A BVD vs Fluke SL935-10k, 0.5 mA 11881 days

Fluke 742A-10 kΩ model, S/N 6250018

Fluke 742A-10k Value Unit
Nominal value 10000 Ω
Operation temperature range +18 … +28 °C
Maximum voltage 30 V
Maximum current 3 mA

This unit belongs to TheSteve’s lab in Canada and was acquired from secondary market. It’s been a part of CalFest 2025 intercomparisons xDevs HQ and TheSteve.

It was shipped with standard courier delivery service in Pelican 1535 hard plastic case together with temperature datalogger and Keithley 2002 DMM.

Standard arrived in clean condition without any exterior damages.

It has sticker on top “10.000037 kΩ” but without date or temperature it’s not very meaningful for metrology purposes.

Back has all typical Fluke factory stickers with serial number in both barcode and human readable forms, PSL certification label D3176 dated 2 November, 1999. Standard was placed in controlled TEC airbath together with other resistance standards. It was allowed to settle at +23.0 °C ±0.05 °C for a week and then measured with MI 6000A resistance bridge for initial points.

These labels also tell us factory TCR values. Calibration value at factory assigned value 10000.024 Ω with TCR coefficients α = +0.030 µΩ/Ω/°C ; β = -0.015 µΩ/Ω/°C2. Temperature stability was remeasured with golden Keysight 3458A DMM during long 3 day 18 hour sweep in programmable air-bath with slow temperature ramps between +18 °C and +28 °C.

Calculated TCR coefficients α = -0.004 µΩ/Ω/°C ; β = -0.012 µΩ/Ω/°C2 with zero TC point at +22.8 °C which is very close to ideal +23.0 °C expectation. Hysteresis after sweep was less than 0.1 µΩ/Ω which is within noise and uncertainty of 3458A measurement.

Model and serial Calibration date Assigned value Uncertainty Calculated drift Notes Age
742A-10k 6250018 2.NOV.1999, factory 10000.024 Ω ± 1.0 µΩ/Ω Ref Original factory label Ref
742A-10k 6250018 30.OCT.2025 by xDevs 10000.09275 Ω ± 0.22 µΩ/Ω +6.875 µΩ/Ω; +0.264 µΩ/Ω/year MIL 6000A vs xDevs ESI SR104, 0.316 mA 9494 days
742A-10k 6250018 31.OCT.2025 by xDevs 10000.09270 Ω ± 0.22 µΩ/Ω +6.870 µΩ/Ω; +0.264 µΩ/Ω/year MIL 6000A vs xDevs ESI SR104, 0.316 mA 9495 days
742A-10k 6250018 1.NOV.2025 by xDevs 10000.09254 Ω ± 0.22 µΩ/Ω +6.854 µΩ/Ω; +0.263 µΩ/Ω/year MIL 6000A vs xDevs ESI SR104, 0.316 mA 9496 days
742A-10k 6250018 30.OCT.2025 by xDevs 10000.09363 Ω ± 0.22 µΩ/Ω +6.963 µΩ/Ω; +0.267 µΩ/Ω/year MIL 6000A vs Fluke SL935-10k, 0.316 mA 9494 days
742A-10k 6250018 31.OCT.2025 by xDevs 10000.09476 Ω ± 0.22 µΩ/Ω +7.076 µΩ/Ω; +0.272 µΩ/Ω/year MIL 6000A vs Fluke SL935-10k, 0.316 mA 9495 days
742A-10k 6250018 1.NOV.2025 by xDevs 10000.09453 Ω ± 0.22 µΩ/Ω +7.053 µΩ/Ω; +0.271 µΩ/Ω/year MIL 6000A vs Fluke SL935-10k, 0.316 mA 9496 days
742A-10k 6250018 26.JAN.2026 by xDevs 10000.09394 Ω ± 0.2 µΩ/Ω +6.994 µΩ/Ω; +0.266 µΩ/Ω/year MIL 6000A vs Fluke SL935-10k, 0.316 mA 9582 days
742A-10k 6250018 2.FEB.2026 by xDevs 10000.09337 Ω ± 0.2 µΩ/Ω +6.937 µΩ/Ω; +0.264 µΩ/Ω/year MIL 6000A vs Fluke SL935-10k, 0.316 mA 9589 days
742A-10k 6250018 28.MAR.2026 by xDevs 10000.09331 Ω ± 0.2 µΩ/Ω +6.931 µΩ/Ω; +0.262 µΩ/Ω/year MIL 6000A vs Fluke SL935-10k, 0.316 mA 9644 days

Overall long-term drift of this Fluke 742A-10kΩ S/N 6250018 estimated at +0.2643 µΩ/Ω/year rate, based on measurements since 30 October 2025 until 28 March 2026.

Fluke 742A-10 kΩ model, S/N 7633009

Fluke 742A-10k 7633009 Value Unit
Nominal value 10000 Ω
Operation temperature range +18 … +28 °C
Maximum voltage 30 V
Maximum current 0.003 A

According to labels this standard was manufactured and calibrated by Fluke lab on February 5, 2001. We have additional calibrations sticker with values from from IRSST . Label stated test current 300 mA instead of 300 µA which I’m sure is just a blatant typo.

Looking at the back we can find pretty impressive low temperature coefficient.

Model and serial Calibration date Assigned value Uncertainty Calculated drift Points in calculation Age
742A-10k 7633009 5.FEB.2001 10000.000 Ω ± 1.00 µΩ/Ω Ref Ref
742A-10k 7633009 24.OCT.2004 10000.007 Ω ± 1.2 µΩ/Ω +0.7 µΩ/Ω, +0.188 µΩ/Ω/a 1357 days
742A-10k 7633009 27.MAR.2022 by Datron 1281 100000.036 Ω ± 1.3 µΩ/Ω +3.6 µΩ/Ω, +0.170 µΩ/Ω/a 7720 days
742A-10k 7633009 28.AUG.2024 by xDevs.com 10000.0328 Ω ± 0.2 µΩ/Ω +3.28 µΩ/Ω, +0.139 µΩ/Ω/a 160 by MI 6000A BVD, 0.5mA 8605 days
742A-10k 7633009 28.NOV.2025 by xDevs.com 10000.0381 Ω ± 0.2 µΩ/Ω +3.81 µΩ/Ω, +0.153 µΩ/Ω/a 160 by MI 6000A BVD, 0.5mA 9062 days
742A-10k 7633009 9.APR.2026 by xDevs.com 10000.0392 Ω ± 0.2 µΩ/Ω +3.92 µΩ/Ω, +0.156 µΩ/Ω/a 160 by MI 6000A BVD, 0.316mA 9194 days

All calibrations performed at +23.0 °C air-bath chamber. No TCR measurements performed at this time for this standard.

Fluke 742A-10 kΩ model, S/N 7815007

Fluke 742A-10k 7815007 Value Unit
Nominal value 10000 Ω
Operation temperature range +18 … +28 °C
Maximum voltage 30 V
Maximum current 0.003 A

According to labels this standard was manufactured and calibrated by Fluke lab on April 26, 2001. We have additional calibrations sticker with values from from IRSST . Label stated test current 300 mA instead of 300 µA which I’m sure is just a blatant typo.

Looking at the back we can find typical tempco values at the date of calibration in Fluke factory.

Model and serial Calibration date Assigned value Uncertainty Calculated drift Points in calculation Age
742A-10k 7815007 26.APR.2001 10000.001 Ω ± 1.00 µΩ/Ω Ref Ref
742A-10k 7815007 24.OCT.2004 10000.007 Ω ± 1.2 µΩ/Ω +0.6 µΩ/Ω, +0.172 µΩ/Ω/a 1277 days
742A-10k 7815007 27.MAR.2022 by Datron 1281 10000.0510 Ω ± 1.3 µΩ/Ω +5.0 µΩ/Ω, +0.239 µΩ/Ω/a 7640 days
742A-10k 7815007 10.JUL.2022 by xDevs.com 10000.0480 Ω ± 0.32 µΩ/Ω +4.7 µΩ/Ω, +0.222 µΩ/Ω/a MI 6010B DCC, 0.316mA 7745 days
742A-10k 7815007 7.APR.2023 by xDevs.com 10000.0530 Ω ± 0.2 µΩ/Ω +5.2 µΩ/Ω, +0.237 µΩ/Ω/a 128 by MI 6000A BVD, 0.5mA 8016 days
742A-10k 7815007 28.AUG.2024 by xDevs.com 10000.0520 Ω ± 0.2 µΩ/Ω +5.1 µΩ/Ω, +0.218 µΩ/Ω/a 160 by MI 6000A BVD, 0.5mA 8525 days
742A-10k 7815007 30.MAR.2026 by xDevs.com 10000.0581 Ω ± 0.2 µΩ/Ω +5.71 µΩ/Ω, +0.229 µΩ/Ω/a 90 by MI 6000A BVD, 0.316mA 9104 days
742A-10k 7815007 3.APR.2026 by xDevs.com 10000.0580 Ω ± 0.2 µΩ/Ω +5.71 µΩ/Ω, +0.229 µΩ/Ω/a 90 by MI 6000A BVD, 0.316mA 9108 days

This standard was tested on 28.AUG.2024 against 742A-10k 7633009 as well with measured ratio between the two 1.00000192 and standard uncertainty ±0.05 ppm. It was also recalibrated end of March 2026.

All calibrations performed at +23.0 °C air-bath chamber. No TCR measurements performed at this time for this standard.

Fluke 742A-19 kΩ model, S/N 6935001

Fluke 742A-19k Value Unit
Nominal value 10000 Ω
Operation temperature range +18 … +28 °C
Maximum voltage 28.5 V
Maximum current 1.5 mA

This unit belongs to xDevs-FL and was acquired from secondary market. It participated in CalFest 2021 intercomparison between xDevs HQ and FL with long-scale DMM.

Unit came with all good labels and has TCR with value. Calibration value was 19000.08 Ω with TCR coefficients α = -0.130 µΩ/Ω/°C ; β = -0.025 µΩ/Ω/°C2. Similar to 1.9 Ω this unit with Verizon sticker but without actual newer value available.

Model and serial Calibration date Assigned value Uncertainty Calculated drift Notes Age
742A-19k 6935001 ? factory 19000.08 Ω ± 1.5 µΩ/Ω Ref Original factory label Ref
742A-19k 6935001 24.NOV.2021 by xDevs 19000.1025 Ω ± 0.8 µΩ/Ω +1.184 µΩ/Ω Transfer from SR104 with Datron 1281 0 days
742A-19k 6935001 12.APR.2023 by xDevs 19000.1051 Ω ± 0.3 µΩ/Ω +1.321 µΩ/Ω, +0.099 µΩ/Ω/a 6000A BVD vs Fluke 742A-10k, 20V 504 days

Fluke 742A-100 kΩ model, S/N 7810004

Fluke 742A-100k Value Unit
Nominal value 100 kΩ
Operation temperature range +18 … +28 °C
Maximum voltage 40 V
Maximum current 1 mA

This unit belongs to xDevs-FL and was acquired from secondary market. It started participation in CalFest intercomparisons from 2022 between xDevs HQ and FL with MI 6000A BVD bridge system.

Unit came without value label but has TCR and serial number. TCR coefficients very good α = -0.010 µΩ/Ω/°C ; β = -0.012 µΩ/Ω/°C2.

Model and serial Calibration date Assigned value Uncertainty Calculated drift Notes Age
742A-100k 7810004 13.JUL.2022 by xDevs 100.000983 kΩ ± 0.3 µΩ/Ω REF µΩ/Ω, REF µΩ/Ω/a 6000A BVD vs ESI SR104, 20V REF days
742A-100k 7810004 12.APR.2023 by xDevs 100.000995 kΩ ± 0.4 µΩ/Ω +0.12 µΩ/Ω, +0.16 µΩ/Ω/a 6000A BVD vs LC-10KA, 20V 274 days

Fluke 742A-1 MΩ model, S/N 5800002

Fluke 742A-1M Value Unit
Nominal value 1 <Ω
Operation temperature range +18 … +28 °C
Maximum voltage 100 V
Maximum current 100 µA

This unit belongs to xDevs-FL and was acquired from secondary market. It started participation in CalFest intercomparisons from 2022 between xDevs HQ and FL with MI 6000A BVD bridge system.

Unit came without value label but has TCR and serial number. TCR coefficients α = +0.170 µΩ/Ω/°C ; β = -0.014 µΩ/Ω/°C2.

Model and serial Calibration date Assigned value Uncertainty Calculated drift Notes Age
742A-1M 5800002 14.JUL.2022 by xDevs 1.00000608 MΩ ± 0.4 µΩ/Ω REF 6000A BVD, 20V REF
742A-1M 5800002 12.APR.2023 by xDevs 1.00000583 MΩ ± 0.5 µΩ/Ω -0.25 µΩ/Ω, -0.33 µΩ/Ω/a 6000A BVD, 20V 274 days

Fluke 742A-10 MΩ model, S/N 5095049

This unit belongs to xDevs-FL and was acquired from secondary market. It started participation in CalFest intercomparisons from 2022 between xDevs HQ and FL with MI 6000A BVD bridge system.

Fluke 742A-10M Value Unit
Nominal value 10 MΩ
Operation temperature range +18 … +28 °C
Maximum voltage 200 V
Maximum current 20 µA

Unlike lower resistance standards Fluke opted to provide only two-terminal outputs on model 742A above 190 kΩ nominal. This is becaouse parasitic contact and cable resistances play very minor role per typical use cases with these higher resistances. Uncertainty of connections would be nearly invisible with typical DMMs, calibrators and bridge systems. Originally this unit was calibrated by Process Instruments in 2017. Unit was aquired without value label but has TCR and serial number. TCR coefficients α = +0.020 µΩ/Ω/°C ; β = +0.064 µΩ/Ω/°C2.

This standard was used for xDevs high-resistance transfers and experiments in recent years. To ensure good accuracy it was shipped to Measurements International calibration laboratory for accredited calibration with low uncertainty. Calibration was performed using MI 6000 binary voltage divider bridge with 20V level, at fixed +23.0 °C air-bath temperature.

Per calibration report this standard was returned with new assigned value 10.000091 MΩ and uncertainty ±0.64 µΩ/Ω. After calibration it was used during xDevs.com CalFest 2022 experiments until return to xDevs FL homelab at CalFest 2023.

Summary of history on this standard available in table as usual. Date of manufacture or factory value from Fluke is unknown because original label was missing.

Model and serial Calibration date Assigned value Uncertainty Calculated drift Notes Age
742A-10M 6715004 3.NOV.2017 by PI 10.000122 MΩ ± 3.80 µΩ/Ω Ref Ref
742A-10M 6715004 1.JUL.2022 by MI 10.000091 MΩ ± 0.64 µΩ/Ω -3.1 µΩ/Ω, -0.66 µΩ/Ω/a 1701 days
742A-10M 6715004 15.JUL.2022 by xDevs 10.0000921 MΩ ± 0.8 µΩ/Ω -3.0 µΩ/Ω, -0.64 µΩ/Ω/a 256 samples via xDevs BVD, 20V 1715 days
742A-10M 6715004 12.APR.2023 by xDevs 10.0001009 MΩ ± 0.8 µΩ/Ω -2.11 µΩ/Ω, -0.388 µΩ/Ω/a 6000A BVD, 20V 1986 days

Fluke 742A-19 MΩ model, S/N 9103002

Fluke 742A-19M Value Unit
Nominal value 19 MΩ
Operation temperature range +18 … +28 °C
Maximum voltage 190 V
Maximum current 10 µA

This unit belongs to xDevs-NLab and was received from eBay, without any previous history or factory labels with original resistance value calibration and TCR. Date of manufacture is unknown but it is newer model without older glorious metal Fluke badge.

It was received stripped of all factory labels except serial number. This means it’s factory TCR and measured value is unknown.

Model and serial Calibration date Assigned value Uncertainty Calculated drift Notes Age
742A-19M 9103002 18.JUL.2022 by xDevs 19.0001510 MΩ ± 1.2 µΩ/Ω Ref 64 samples via xDevs BVD, 20V Ref

Fluke 742A-19 MΩ model, S/N 6004003

Fluke 742A-19M Value Unit
Nominal value 19 MΩ
Operation temperature range +18 … +28 °C
Maximum voltage 190 V
Maximum current 10 µA

This unit belongs to xDevs-FL and was acquired from secondary market. It started participation in CalFest intercomparisons from 2022 between xDevs HQ and FL with MI 6000A BVD bridge system.

Unit came with value labels and TCR. Calibration value was 19.00009 Ω with TCR coefficients α = +0.120 µΩ/Ω/°C ; β = -0.040 µΩ/Ω/°C2.

Model and serial Calibration date Assigned value Uncertainty Calculated drift Notes Age
742A-19M 6004003 Unknown 19.00009 MΩ ± 20 µΩ/Ω ? Fluke ?
742A-19M 6004003 15.JUL.2022 by xDevs 19.000182 MΩ ± 1.7 µΩ/Ω REF 256 samples via xDevs BVD, 20V REF
742A-19M 6004003 12.APR.2023 by xDevs 19.0001983 MΩ ± 1.8 µΩ/Ω +0.858 µΩ/Ω, +1.16 µΩ/Ω/a 6000A BVD, 20V 271 days

Measuring self-heating effect of Fluke 742A 1 Ω resistors and fixed-point TCR curves

Standard resistors can be calibrated at different currents. The main thesis of this section is that measurement current is heating the internal resistance element of the resistor and that heat, along with measurement time, is the primary cause for difference in calibration values at different currents. This is especially critical for high-power resistance standards and low-ohmic shunts, but also affecting even “normal” 1 Ω – 1000 Ω decade value resistance standards used to calibrate long-scale DMMs, calibrators, current and voltage sources.

During experiment, we establish TCR curves for two Fluke 742A-1 resistors, including theoretical 0 mW power curve.

Used equipment for this study:

DataProof scanner used to connect either resistor A or resistor B to RS terminals of the DCC bridge. Resistors are placed in TEC1 chamber at programmatically controlled temperatures (+19 °C to +27 °C). Resistor C is connected to RX terminals of the bridge. Resistor C resides in TEC2 chamber at constant +23.2 °C.

Theory

TCR of standard resistors are defined as quadratic function with α and β coefficients. We will call a “HAT” the flattest portion of the TCR curve, and T_HAT and R_HAT would be the temperature and resistance values corresponding to this flattest point.

Our 742A-10 resistor will be used at 1mW and 0.5 mW due to 10:1 RX to RS ratio. This minimizes its self-heating comparing to usual 10mW.

Additionally, it’s T_HAT at point +24.5 °C so it has the flattest section of its TCR between +23 °C and +25 °C and therefore even with some small amount of internal heating its resistance should be very stable.

Nominally Fluke 742A-1 resistors are calibrated at 10 mW power (100 mA). We will conduct all experiments with 10mW power and also half power 5mW (70.7 mA) to derive the effects of power heating.

Due to it’s internal construction Fluke 742A-1 resistors have significant thermal isolation between their enclosure and internal resistive elements. Due to this design, it takes a very long time for the temperature difference to sink in or self-heating to equalize and stabilize making precise experiments with Fluke 742A relatively time consuming. At the same time this thermal isolation also helps to keep resistance stable, even with small fluctuations in ambient temperature during normal use.

After settling TEC1 air-bath to set temperature and letting resistors to soak for about 10 hours, we will measure resistors interchangeably Res A for 10 hours, Res B for 10 hours, Res A for 10 hours and again Res B for 10 hours. Since we are constantly measuring against RX, the resistor C will always be at constant current i.e. constant internal temperature and also constant external temperature of TEC2 (+23.2 °C).

In each cycle we will observe how resistors A and B go from being cold i.e. internal temperature being equal to chamber TEC1 temperature, to being measured and heated by the measurement current. While one resistor is being measured the other resistor is cooling down back to chamber temperature. Repeatability of measurements using transfer 10 Ω standard was confirmed by multiple measurements with same settings.

This process allows us to see how many hours it takes for a measurement to reach its terminal value, as well as values within ±0.02 µΩ/Ω and ±0.05 µΩ/Ω of the final value.

There experiments are done at +19 °C, +21 °C, +23 °C, +25 °C and +27 °C temperatures. While just 3 temperature points are sufficient to produce TCR α and β coefficients, recording 4 or 5 points helps with additional data cross checking and confidence.

After running all measurements at 100 mA test current we will repeat them at 70.7 mA. For Resistor A recorded values in Ohms are as follows:

TEMP, °C Current 100 mA Current 70.7 mA
+27 °C 1.00004298 Ω 1.00004308 Ω
+25 °C 1.00004387 Ω
+23 °C 1.00004454 Ω 1.00004459 Ω
+21 °C 1.00004492 Ω
+19 °C 1.00004507 Ω 1.00004508 Ω

This data available in graphical and Excel-file format:

For resistance standard B similar data sequence was collected:

TEMP, °C Current 100 mA Current 70.7 mA
+27 °C 1.00000764 Ω 1.00000770 Ω
+23 °C 1.00000870 Ω 1.00000877 Ω
+21 °C 1.00000912 Ω
+19 °C 1.00000958 Ω 1.00000964 Ω

This data available in graphical and Excel-file format:

Earlier initial calibration obtained from MIL 6010B bridge also shown with cyan diamond at fixed +23.0 °C resistor body temperature.

Simple formula calculation from Fluke 742A manual produces new calculated α and β:

Resistance standard Test current, A α coefficient, µΩ/Ω/°C β coefficient, µΩ/Ω/°C2
Unit A, S/N 6270023 0.1000 -0.261 -0.032
Unit A, S/N 6270023 0.0707 -0.250 -0.032
Unit A, optimized values 0.1000 -0.261 -0.032
Unit B, S/N 7681001 0.1000 -0.243 -0.006
Unit B, S/N 7681001 0.0707 -0.242 -0.006
Unit B, optimized values 0.1000 -0.240 -0.006

Please note that measured β is identical to manufacturers provided β of -0.032µΩ/Ω, while both resistance value at +23 °C and α changed over years since resistor was manufactured.

Given α and β R_HAT or the highest RS resistance point regardless of temperature was calculated to be 1.00004508 Ω at 100 mA and 1.00004509 Ω at 70.7 mA or only 0.009 µΩ/Ω different. (formulas for calculation are in Excel).

This corresponds to difference in resistor C resistance at 1 mW and 0.5 mW and could be considered negligible as it is comparable to the used DCC bridge noise and repeatability. Therefore, we consider resistance C to be constant for this experimentation.

T_HAT was calculated to be +18.93 °C at 100 mA and +19.10 °C at 70.7 mA. Considering that β coefficient are the same for 100 mA and 70.7 mA, the major effect of double the heating comparing to the 70.7 mA resulted in “pushing” the TCR curve sideways by 0.17 °C. In other words, doubling the power from 5 mW to 10 mW resulted in additional heating of 0.17 °C. The zero power curve is produced by “pushing” 5 mW TCR 0.17 °C in opposite direction.

Conclusions for TCR and power effects in Fluke 742A resistors.

100 mA in 742A-1 resistor A produced 0.34 °C of self-heating that should be considered as additional to the ambient temperature of the resistor.

From data analysis at 100 mA of current it takes 742A-1A approximately 6 hours to reach its terminal value. 3 hours are sufficient for 0.02 µΩ/Ω stability window and 1 hour for 0.05 µΩ/Ω stability window.

At 70.7 mA the corresponding times were 5 hours, 2.5 hours and 0.5 hours respectively.

When resistor is measured at 100 mA very quickly, its value initially corresponds to theoretical 0 mA TCR curve, after 6 hours of measurement it will be at 100 mA TCR curve and for intermediate times it will be in the middle between these two curves.

Specifically for resistor A used in this article: at +28 °C there might be as much as 0.2 µΩ/Ω difference in R28 value, depending on measurement current (100 mA or less) and measurement time. At +23 °C, difference diminishes to 0.1 µΩ/Ω and at +19 °C it is completely negligible at 0.02 µΩ/Ω.

Specifically for resistor B, 100 mA produced approximately 0.5 °C of heating. β was also identical to manufacturer’s data and R23 and α both drifted.

The resistance value difference between different measurement currents (under 100 mA) and measurement times were 0.13 µΩ/Ω or less regardless of the temperature (+18 °C to +28 °C).

Long-term drift history of Fluke 742A standards

Summary table with all long-term drift estimations for the Fluke 742A standards presented below.

Model and serial Age Last measured value Calculated drift Points in calculation
742A-1 5825005 30 years 1.00003702 Ω ±0.25 µΩ/Ω +1.22 µΩ/Ω/year (+1.14 since previous calibration in 2011) Jun 1993, Jun 2011, Apr 2022, Apr 2023, Aug 2023
742A-1 6270023 22 years 1.00004464 Ω ±0.33 µΩ/Ω +1.4 µΩ/Ω/year (+0.65 since previous calibration in 2020) Nov 1999, Oct 2020, Mar 2022
742A-10k 5820007 29 years 10000.08865 Ω ±0.2 µΩ/Ω +0.285 µΩ/Ω/year (+0.17 since previous calibration in 2011) Jun 1993, Jun 2011, Apr 2022, Apr 2023, Jan 2024

Keep in mind that these numbers are unique to particular reference. They should meet the typical behaviour and Fluke specifications, but actual drift rate for each unit depends on many real world parameters such as storage conditions, test currents, usage frequency, connector wear and tear, etc.

xDevs.com transfer reports with 742A results

Traceability calibration map is available for xDevs transfers on 742A standards, obtained during April 2023 meet-up.

xDevs.com CalFest 2025 results with 742A from 1 Ω to 10000 Ω

Database of labels/results from other 742A from public information

Model and serial Factory value Last measured value α β Reference
742A-1 4596007 0.9999995 Ω 0.9999837 Ω ±0.5-6,k=3, 50mA, +20°C, 27.JUL.2018 by SIMT 0.160-6 -0.068-6 bbs.38hot.net
742A-1 4935018 1.0000467 Ω Pressure coefficient -0.17-9/hPa, Current coefficient -0.0131 µΩ/Ω/10mA -0.08-6 -0.044-6 AFRIMETS.EM-S1
742A-1 5435027 1.0000004 Ω ? 0.069-6 -0.042-6 etoysbox.jp
742A-1 5465010 1.0000011 Ω ? -0.073-6 -0.051-6 eBay listing
742A-1 5715009 1.0000005 Ω ? -0.040-6 -0.043-6 eBay listing
742A-1 5825005 1.0000002 Ω 1.00003626 Ω ±0.4-6,k=2, 100mA, +23.0 °C, 12.APR.2023 by DCC, +1.2 µΩ/Ω/year -0.060-6 (-0.109-6 current) -0.037-6 (-0.040-6 current) xDevs unit
742A-1 5835001 1.0000010 Ω -0.110-6 -0.044-6 eBay listing
742A-1 5920006 1.0000001 Ω -0.110-6 -0.038-6 eBay listing
742A-1 6270023 1.0000131 Ω 1.00004374 Ω ±0.46-6,k=2, 100mA, +23.0 °C, 27.OCT.2020 by IET -0.02-6 -0.032-6
742A-1 6590026 1.0000002 Ω Fluke PSL DR569-26, 16 September 1996 +0.00-6 -0.04-6 eBay listing
742A-1 6855020 0.99999680 Ω -0.14-6 -0.029-6 eBay listing
742A-1 6915029 1.00002333 Ω ±0.17-6,k=2, 100mA, +23.0 °C, 12.DEC.2018 by PI -0.087-6 -0.0219-6 eBay listing
742A-1 7681001 0.9999995 Ω Fluke 1R0199 05.FEB.2001 -0.160-6 -0.007-6
742A-1 7790003 No data No data -0.044-6 -0.011-6 Tom
742A-1 7791006 0.9999991 Ω Fluke 1R0331 09.MAY.2001 -0.007-6 -0.028-6 eBay listing
742A-1 7792005 NA Ω 0.99999955 – 0.99999968 Ω NIMT & NMIA DCC -0.063-6 -0.013-6 KCDB APMP.EM-S6
742A-1 8140001 1.0000018 Ω 16-Jul-2002, Fluke PSL label -0.015-6 -0.018-6 Fluke 5700-7002 ACAL kit
742A-1 8927002 1.0000000 Ω -0.09-6 -0.036-6 Reps
742A-1 9553013 1 Ω ? 0.119-6 -0.038-6 eBay listing
742A-1.9 5325001 1.899997 Ω ? 0.060-6 -0.047-6 eBay listing
742A-1.9 9558011 1.9 Ω ? 0.054-6 -0.029-6 eBay listing
742A-10 4920006 10.000470 Ω Pressure coefficient -0.178-9/hPa, Current coefficient 0.0027 µΩ/Ω/1mA -0.02-6 -0.036-6 AFRIMETS.EM-S1
742A-10 5290004 10.000000 Ω Fluke PSL, 1 Aug, 1991 0.010-6 -0.058-6 eBay listing
742A-10 5320005 10.000003 Ω Fluke PSL, 1991 0.090-6 -0.062-6 eBay listing
742A-10 5360007 10.000003 Ω Fluke PSL, 21 October 1991, Measured 10.0001012 Ω January 25, 2023 0.050-6 -0.066-6 eBay listing
742A-10 7170003 NA Ω 10.000121 Ω 0.110-6 -0.038-6 eBay auction
742A-10 9558012 10 Ω ? -0.037-6 -0.045-6 eBay listing
742A-25 9566001 25 Ω ? 0.106-6 -0.048-6 eBay listing
742A-100 5485007 100.00682 Ω Pressure coefficient -0.185-9/hPa, Current coefficient 0.0013 µΩ/Ω/0.1mA -0.009-6 -0.044-6 AFRIMETS.EM-S1
742A-100 6560003 100 Ω 100.001125 Ω ±0.95 µΩ/Ω 0.040-6 -0.044-6 eBay listing
742A-100 7798002 NA Ω 100.000215 – 100.000254 Ω NIMT & NMIA DCC -0.029-6 -0.023-6 KCDB APMP.EM-S6
742A-100 9566002 100 Ω ? 0.003-6 -0.006-6 eBay listing
742A-1k 5505003 1000 Ω 1000.0613 Ω ±0.3 µΩ/Ω, 11/24/2021, 3.16mA 0.050-6 -0.018-6 MMLab
742A-1k 5475001 1000.0001 Ω Fluke PSL, 18 Mar, 1992 -0.006-6 -0.015-6 eBay listing
742A-1k 5335001 1000.0005 Ω Fluke PSL, 29 Aug, 1991 0.020-6 -0.020-6 eBay listing
742A-1k 5475001 1000.0001 Ω Fluke PSL, 18 Mar, 1992 -0.006-6 -0.015-6 eBay listing
742A-1k 7733002 1000.006 Ω Pressure coefficient -0.185-9/hPa, Current coefficient 0.0012 µΩ/Ω/0.1mA 0.05-6 -0.022-6 AFRIMETS.EM-S1
742A-1k 9566003 1000 Omega; ? -0.031-6 -0.016-6 eBay listing
742A-10k TS48981 NA Ω 10000.2234 – 10000.2279 Ω NIMT & NMIA DCC -0.08-6 -0.020-6 KCDB APMP.EM-S6
742A-10k 3316003 10000 Ω ? 0.126-6 -0.009-6 eBay listing
742A-10k 4690001 10000.4403 Ω ±0.33-6,k=2, 0.3mA, +23.0 °C, 12.DEC.2018 by PI -0.01-6 -0.034-6 eBay listing
742A-10k 5065038 10000.134 Ω Pressure coefficient 0.148-9/hPa, Current coefficient -0.0647 µΩ/Ω/0.01mA +0.04-6 -0.020-6 AFRIMETS.EM-S1
742A-10k 5190011 9999.991 Ω Fluke PSL DR243-09, 26 Mar 1991, current value 10000.0916 Ω +0.031-6 -0.032-6 from DN Lab
742A-10k 5820007 10000.006 Ω Current value is 10000.0904 Ω on 25.MAY.2024 0.000-6 -0.024-6 xDevs.com unit
742A-10k 5340002 10000.003 Ω ? 0.000-6 -0.012-6 eBay listing
742A-10k 5380009 10000.002 Ω Fluke PSL DR243-09, 26 November 1991 -0.080-6 -0.020-6 eBay listing
742A-10k 5435022 9999.997 Ω ? 0.050-6 -0.010-6 etoysbox.jp
742A-10k 5565006 9999.996 Ω ? +0.120-6 -0.049-6 EEVBlog forum
742A-10k 5680003 10000.004 Ω ? -0.020-6 -0.027-6 Listing
742A-10k 5925007 9999.997 Ω Fluke PSL DR432-07, 7 December 1993 -0.010-6 -0.010-6 Listing
742A-10k 6350011 10000.004 Ω ? 0.030-6 -0.025-6 bbs.38hot.net
742A-10k 6595008 9999.995 Ω Fluke PSL DR572-08, 1 August 1996 +0.03-6 -0.012-6 eBay listing
742A-10k 7240002 10.000038 kΩ 0.04-6 0.012-6 eBay listing
742A-10k 7633009 10000.000 Ω Fluke 1R0129 05.FEB.2001 0.000-6 -0.006-6 Listing
742A-10k 7694007 10000.003 Ω ? -0.010-6 -0.006-6 EEVBlog forum
742A-10k 7655004 No data No data +0.110-6 -0.016-6 Tom
742A-10k 7815006 9999.999 Ω ? 0.000-6 -0.032-6 Reps
742A-10k 7815007 10000.001 Ω Fluke 1R0403 26.APR.2001 0.105-6 0.000-6 Listing
742A-10k 8153001 9999.990 Ω 17-Jul-2002, Fluke PSL label 0.069-6 -0.011-6 Fluke 5700-7002 ACAL kit
742A-10k 9525002 10 kΩ ? 0.039-6 0.004-6 eBay listing
742A-10k ? 10 kΩ Fluke PSL DR446-05, 19.MAR.1994 0.000-6 -0.028-6 Auction listing
742A-19k 6375001 19.000020 kΩ Fluke PSL DR528-01, 11.SEP.1995 -0.120-6 -0.000-6 eBay listing
742A-19k 7852001 19.000015 kΩ ? 0.076-6 -0.018-6 eBay listing
742A-19k 9558013 19 kΩ ? 0.033-6 -0.017-6 eBay listing
742A-100k 3316004 100 kΩ ? -0.008-6 0.002-6 eBay listing
742A-100k 8363002 NA kΩ 100.00023 – 100.00034 kΩ NIMT & NMIA DCC 0.134-6 -0.004-6 KCDB APMP.EM-S6
742A-100k 9558014 100 kΩ ? 0.055-6 -0.014-6 eBay listing
742A-1M 3316004 1 MΩ ? 0.049-6 -0.018-6 eBay listing
742A-1M 8363001 NA MΩ 0.9999998 – 1.0000010 MΩ NIMT & NMIA DCC 0.014-6 -0.013-6 KCDB APMP.EM-S6
742A-1M 9558009 1 MΩ ? 0.074-6 0.005-6 eBay listing
742A-10M 3291001 10 MΩ ? 0.257-6 -0.015-6 eBay listing
742A-10M 9558007 10 MΩ ? 0.121-6 -0.036-6 eBay listing
742A-19M 5455005 19.00000 MΩ Fluke PSL DR282-05, 3.FEB.1992 0.170-6 -0.250-6 eBay listing
742A-19M 6460005 19.00003 MΩ ? 0.110-6 -0.086-6 eBay listing
742A-19M 6480001 18.99995 MΩ Fluke PSL DR550-01, 18.DEC.1995 0.060-6 -0.052-6 eBay listing
742A-19M 9558008 19 MΩ ? 0.365-6 -0.034-6 eBay listing
742A-19M 9103002 19 MΩ ? ? ? eBay listing

The author would like to express our thanks to the Measurements International for excellent calibration services on our primary 10000 Ω ESI SR104 standards and to Nikonoid for temperature/power effect testing, analysis and cross-verification of data sets. Also special thanks goes to long-term xDevs.com supporter Todd M. for participation in interlab comparisons for resistance and voltage. Also author express appreciation to DN Lab for providing Fluke 742A-10k unit for study.

Discussion is very welcome thru comment section or at our own IRC chat server: irc.xdevs.com (port 4808, channel: #xDevs.com) or at our xDevs forum. If you have information and performance reports of Fluke 742A standards not mentioned or listed in this article, feel free to provide them and we will add them with next article update. We are looking forward to add another 742A units in dataset as it becomes available.

Projects like this are born from passion and a desire to share how things work. Education is the foundation of a healthy society - especially important in today's volatile world. xDevs began as a personal project notepad in Kherson, Ukraine back in 2008 and has grown with support of passionate readers just like you. There are no (and never will be) any ads, sponsors or shareholders behind xDevs.com, just a commitment to inspire and help learning. If you are in a position to help others like us, please consider supporting xDevs.com’s home-country Ukraine in its defense of freedom to speak, freedom to live in peace and freedom to choose their way. You can use official site to support Ukraine – United24 or Help99. Every cent counts.

Author: Ilya Tsemenko
Created: Sept. 15, 2020, 6:04 a.m.
Modified: Aug. 8, 2026, 2:44 a.m.