Fluke 732 series DC Voltage Standards service, modifications, testing

Contents

Intro

In this article legacy, but still very good Fluke 732A DC Voltage standard will be serviced and tested. We have multiple units so inter-comparisons will be performed to obtain relative drift rate.

Disclaimer

Redistribution and use of this article or any images or files referenced in it, in source and binary forms, with or without modification, are permitted provided that the following conditions are met:

  • Redistributions of article must retain the above copyright notice, this list of conditions, link to this page (https://xdevs.com/fix/f732a/) and the following disclaimer.
  • Redistributions of files in binary form must reproduce the above copyright notice, this list of conditions, link to this page (https://xdevs.com/fix/f732a/), and the following disclaimer in the documentation and/or other materials provided with the distribution, for example Readme file.

All information posted here is hosted just for education purposes and provided AS IS. In no event shall the author, xDevs.com site, or Fluke or any other 3rd party be liable for any special, direct, indirect, or consequential damages or any damages whatsoever resulting from loss of use, data or profits, whether in an action of contract, negligence or other tortuous action, arising out of or in connection with the use or performance of information published here.

If you willing to contribute or have interesting documentation to share regarding pressure measurements or metrology and electronics in general, you can do so by following these simple instructions.

Manuals and service information

Comparison for known specifications of production resistance standards:

Specification Fluke 732A Fluke 732B Fluke 734A Datron 4910 Fluke 5720A calibrator
Output voltage, DC 1V, 1.018V, 10V 1.018V, 10V 4 × 1.018V, 10V 4 × 1.018V, 10V up to 1100V
Accuracy (1y, 10V) ±6 ppm ±2 ppm ±2 ppm ±1 ppm ±2 ppm
Accuracy (30 day, 10V) ±0.5 ppm ±2 ppm ±2 ppm ±1 ppm ±2 ppm
Temperature coefficient, ppm ±0.05 ppm/°C ±3 ppm ±1.5 ppm <±0.1 ppm/°C ±3 ppm max (18-28°C)
Adjustment range 50µV, <0.05ppm
Long term 1 year stability, ppm ? , ? ±8 ppm ±4 ppm ±0.5 ppm ±2 ppm
Reference design solution Motorola SZA263/LTFLU-1 up to 4 x Linear LTZ1000ACH 2 x LTFLU-1
Max load current 12 mA 200 mA 600 µA 10 mA 418 mA
Output noise <1 µVRMS, 0.1-10Hz
Construction type Rugged half-rack module Rugged ¼-rack module Full-width 19” rack frame Rugged half-rack module Full-size rackmount
Active thermal compensation Yes, +45° C oven assembly Yes, multiple sections
Temperature sensor Yes
Operation range, ambient °C +0…+40 0…+50
Power requirements 100,120,220,240VAC or 24-40VDC Not required, passive device
Backup/offline power supply Internal +24V Battery bank, 12 hours life Not required, passive device
Dimensions, weight 191 × 221 × 603 mm, 12.3kg 254 × 206 × 311 mm, 4.8kg 86 × 105 × 127 mm, 730g
MSRP ~$2000 USD $9000 USD $37000+ USD $9000+ USD $30000+ USD

Design and construction

Author: Todd M.
Published: June 1, 2017, 12:01 a.m.
Modified: Oct. 3, 2018, 11:45 a.m.